Hello,
I have the EMF-390. The lowest RF value, no matter where I am, is always 89.8 pw/cm2, even if I wrap the device in aluminium foil. I don't think that's normal. Does the device need to be calibrated or is there something else to adjust? Many thanks in advance. Stephan |
|
Reply #1
EmfDev
    
2314 Posts |
Posted - 04/10/2019 : 12:12:20
|
Hi Stephan, 89.8 pw/cm2 is a really small number. It's equal to 0.0000898 microWatts per cm2 = 0.0000000898 mW/cm2, which is almost 0. So just treat it as 0. |
 |
|
Reply #2
Stephan

Germany
8 Posts |
Posted - 04/11/2019 : 04:50:46
|
Hi, According to technical data, the resolution is 0.01 microWatt/m2. 89.8 pw/cm2 is equal to 0.898 microWatt/m2, if I calculated correctly. This value is clearly above the resolution of 0.01 microWatt/m2 and cannot be assumed to be zero. For bedrooms, for example, < 0.1 microWatt/m2 is recommended by an institute for building biology. With the EMF-390 it is not possible to measure in this range if I have to assume 89.8 pw/cm2 as zero and this despite the given resolution of 0.01 microWatt/m2. Do I see that correctly? Many thanks Stephan
|
 |
|
Reply #3
EmfDev
    
2314 Posts |
Posted - 04/11/2019 : 11:43:48
|
Hi Stephan, I need to check with the design for this. You are correct with the conversion. But do you have reference for the < .1 uW/m2? |
 |
|
Reply #4
Stephan

Germany
8 Posts |
Posted - 04/12/2019 : 00:29:58
|
Hi, You can find the info regarding the reference for bedrooms here: https://www.baubiologie.de/downloads/building-biology-guidelines-english.pdf
Is my conversion really correct, or is it the other way around? Is the power density with the same radiation at 1m2 not lower than at 1cm2? 100pw/cm2 = 0.01pw/m2 = 0.00000001 microWatt/m2 Please check it out. Thank you! Stephan
|
 |
|
Reply #5
EmfDev
    
2314 Posts |
Posted - 04/12/2019 : 14:49:58
|
You were correct, 100pw/cm2 * 100cm/m * 100cm/m = 1000000pw/m2 = 1uW/m2. But the EMF meter should reach <10 pW/cm2 or <.1 uW/m2. It was tested. If not maybe the noise is coming from itself. |
 |
|
Reply #6
Stephan

Germany
8 Posts |
Posted - 04/12/2019 : 22:58:43
|
Yeah, great! Sorry for the confusion I may have created. I am very satisfied with the EMF-390 and will recommend it to others. Thank you! Stephan
|
 |
|
|