GQ Electronics Technical Support Forum Active Users: / Visits Today:
Highest Active Users:
GQ Electronics Technical Support Forum
Home | Profile | Register | Active Topics | Members | Search | FAQ
Username:
Password:
Save Password
Forgot your Password?

 All Forums
 GQ Electronics Forums
 3. GQ EMF EF Meter RF Spectrum Power Analyzer
 Excessive readings in EMF?

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
wolfdogg Posted - 05/29/2024 : 11:11:35
I wanted to check a high voltage line, i think it was a 10k+v transmission line. I have my EMF units set to uT (MicroT). I walked underneath it the powerline and around the field nearby. At first it was doing normal ranges like 1.5xx uT. This went on for a couple minutes. Then as we were walking back, and starting to move (10 feet from being directly below it and onwards) the meter (I had it on vertical mode) the meter spiked, started moving to "XTREME" level, led went red, was reading 600.xxx, then eventually 1000.xxx. Assuming it doesn't change its unit of measure is that 1000 uT? Does it change units of measure? if so that may have been reading 1000mT, or even worse 1000T? to be honest i didn't notice the unit of measure at that point.

From my research that's about 10 times a very high level area and not safe to live nearby.

As we were walking away, not about maybe 40-50+ft from being directly under it that's when it was getting down to about 600.xxx, led was red still, said XTREME.

So at that point, without moving any further, we decided to cycle the power, which then it started reading back to normal 1.5-3 as expected being under the powerline.



I did do a log, and I have that here. its filtered by powerline. NOtice the spike at 44383.266 pw/cm2.

Date and Time	mG	V/m	mW/m2	mW/cm2	pW/cm2	 Possible Source
5/27/2024 12:09	9.1	68.9	0.001	0	94.128	Power Line
5/27/2024 12:09	9.1	68.9	0.001	0	94.128	Power Line
5/27/2024 12:09	11.5	27	0.003	0	308.245	Power Line
5/27/2024 12:10	8.6	90.7	0.001	0	115.822	Power Line
5/27/2024 12:11	5.8	90.7	0.001	0	147.847	Power Line
5/27/2024 12:12	13.5	96.4	0.009	0	925.239	Power Line
5/27/2024 12:12	13.6	80.2	0.002	0	244.363	Power Line
5/27/2024 12:12	13.2	73.6	0.004	0	380.601	Power Line
5/27/2024 12:12	12.1	75.5	0.005	0	546.052	Power Line
5/27/2024 12:12	13.2	74.3	0.002	0	157.697	Power Line
5/27/2024 12:12	12.2	75.7	0.001	0	70.039	Power Line
5/27/2024 12:12	13.2	94.5	0.003	0	323.893	Power Line
5/27/2024 12:12	12.6	69.4	0.005	0	522.293	Power Line
5/27/2024 12:12	12.2	99.2	0.001	0	75.992	Power Line
5/27/2024 12:12	12.7	73.5	0.004	0	440.449	Power Line
5/27/2024 12:12	12.5	86.1	0.003	0	316.039	Power Line
5/27/2024 12:13	13.2	73.2	0.109	0	10880.653	Power Line
5/27/2024 12:14	3	63.1	0.007	0	671.959	Power Line
5/27/2024 12:14	3.5	91.2	0.003	0	273.904	Power Line
5/27/2024 12:14	3.5	91.5	0.002	0	157.697	Power Line
5/27/2024 12:14	3.4	93.1	0.011	0	1055.339	Power Line
5/27/2024 12:14	3.1	90.7	0.001	0	109.564	Power Line
5/27/2024 12:14	2.8	85.5	0.011	0	1145.438	Power Line
5/27/2024 12:14	3.1	81.7	0.013	0	1300.854	Power Line
5/27/2024 12:14	2.8	64.5	0.002	0	184.528	Power Line
5/27/2024 12:16	2.8	302	0.005	0	498.987	Power Line
5/27/2024 12:17	5.3	569	0.009	0	912.772	Power Line
5/27/2024 12:17	4	76.9	0.007	0	677.251	Power Line
5/27/2024 12:19	6	64	0.001	0	64.132	Power Line
5/27/2024 12:19	3.5	89	0.001	0	64.132	Power Line
5/27/2024 13:11	4.6	0	0.444	0	44383.266	Power Line


There were some other higher numbers, and not all were classified as powerline, so i figured i will just show the whole log starting just before it spiked up:

5/27/2024 12:11	5.8	90.7	0.001	0	147.847	Power Line
5/27/2024 12:11	6	288.8	0.001	0	106.453	AC EF
5/27/2024 12:11	5.6	466.4	0.004	0	393.155	AC EF
5/27/2024 12:11	6.8	711	0.005	0	527.008	Mixed
5/27/2024 12:11	7.3	1007	0.009	0	888.123	AC EF
5/27/2024 12:11	8	1001	0.003	0	259.02	AC EF
5/27/2024 12:11	6.7	993.5	0.001	0	70.039	AC EF
5/27/2024 12:11	6.7	987.5	0.001	0	97.191	AC EF
5/27/2024 12:11	7	983.5	0.013	0	1285.505	AC EF
5/27/2024 12:11	5.8	527	0.001	0	100.267	AC EF
5/27/2024 12:11	7	504	0.001	0	118.97	Mixed
5/27/2024 12:11	6.6	1017.5	0.005	0	527.008	AC EF
5/27/2024 12:11	7.1	1006.5	0.001	0	64.132	Mixed
5/27/2024 12:11	7.1	1002.5	0.01	0	976.08	AC EF
5/27/2024 12:11	7.2	990	0.009	0	869.884	Mixed
5/27/2024 12:11	7.7	992.5	0.003	0	285.219	Mixed
5/27/2024 12:11	7	494.6	0.001	0	67.08	Mixed
5/27/2024 12:11	7.3	980	0.006	0	615.064	AC EF
5/27/2024 12:11	9.5	1015.5	0.003	0	259.02	AC EF
5/27/2024 12:11	10.2	1026	0.004	0	355.923	Mixed
5/27/2024 12:11	10.4	1017.5	0.005	0	508.255	AC EF
5/27/2024 12:11	8.6	1009	0.005	0	531.742	AC EF
5/27/2024 12:11	6.9	1009	0.005	0	508.255	Mixed
5/27/2024 12:11	7.3	996	0.001	0	81.991	AC EF
5/27/2024 12:11	6.6	898	0.001	0	64.132	Mixed
5/27/2024 12:11	6	985.5	0.005	0	527.008	Mixed
5/27/2024 12:11	5.5	953.5	0.003	0	289.02	Mixed
5/27/2024 12:11	4.9	996	0.001	0	81.991	AC EF
5/27/2024 12:11	4.7	880.5	0.001	0	67.08	Mixed
5/27/2024 12:11	5.2	997.5	0.001	0	141.343	AC EF
5/27/2024 12:11	5.5	1007.5	0.005	0	503.612	Mixed
5/27/2024 12:12	7.4	569.5	0.005	0	531.742	AC EF
5/27/2024 12:12	5.5	91.7	0.005	0	531.742	Mixed
5/27/2024 12:12	6.9	984.5	0.003	0	259.02	AC EF
5/27/2024 12:12	7.4	985.5	0.18	0	18039.438	AC EF
5/27/2024 12:12	7.3	975.5	0.099	0	9899.191	AC EF
5/27/2024 12:12	9.4	975	2.23	0	222989.469	AC EF
5/27/2024 12:12	10	1017	1.824	0	182401.438	AC EF
5/27/2024 12:12	10.1	563.5	1.852	0	185243.281	Mixed
5/27/2024 12:12	9.4	1017.5	0.005	0	489.789	AC EF
5/27/2024 12:12	8.8	711	0.001	0	144.589	Mixed
5/27/2024 12:12	9.8	858.5	0.001	0	106.453	Mixed
5/27/2024 12:12	10.2	1003.5	0.001	0	78.985	Mixed
5/27/2024 12:12	10.6	408.2	1.065	0	106515.516	AC EF
5/27/2024 12:12	11.7	796.5	1.065	0	106515.516	AC EF
5/27/2024 12:12	12	843.5	0.001	0	78.985	AC EF
5/27/2024 12:12	12	676.5	0.001	0	97.191	AC EF
5/27/2024 12:12	11.2	302	0.001	0	97.191	Mixed
5/27/2024 12:12	12.3	241.4	0.005	0	546.052	AC EF
5/27/2024 12:12	11.9	645.5	0.001	0	94.128	AC EF
5/27/2024 12:12	12.5	137.6	0.005	0	489.789	AC EF
5/27/2024 12:12	13.5	96.4	0.009	0	925.239	Power Line
5/27/2024 12:12	13.6	80.2	0.002	0	244.363	Power Line
5/27/2024 12:12	13.2	73.6	0.004	0	380.601	Power Line
5/27/2024 12:12	12.1	75.5	0.005	0	546.052	Power Line
5/27/2024 12:12	12.3	129.6	0.006	0	575.174	AC EF
5/27/2024 12:12	13.2	74.3	0.002	0	157.697	Power Line
5/27/2024 12:12	13.4	169.6	0.003	0	296.665	AC EF
5/27/2024 12:12	13.1	122.2	0.005	0	453.699	AC EF
5/27/2024 12:12	12.8	171.2	0.003	0	304.37	AC EF
5/27/2024 12:12	13.6	288.8	0.005	0	471.605	AC EF
5/27/2024 12:12	13.5	215.6	0.001	0	67.08	AC EF
5/27/2024 12:12	13.7	179.4	0.01	0	976.08	AC EF
5/27/2024 12:12	12.4	105.2	0.001	0	147.847	AC EF
5/27/2024 12:12	12.2	75.7	0.001	0	70.039	Power Line
5/27/2024 12:12	12.2	786.5	0.006	0	555.685	AC EF
5/27/2024 12:12	12.5	168.4	0.001	0	115.822	AC EF
5/27/2024 12:12	12.9	133.6	0.001	0	85.007	AC EF
5/27/2024 12:12	13.2	94.5	0.003	0	323.893	Power Line
5/27/2024 12:12	12.6	69.4	0.005	0	522.293	Power Line
5/27/2024 12:12	12.2	179.8	0.001	0	112.687	AC EF
5/27/2024 12:12	12.3	168.8	0.002	0	174.369	AC EF
5/27/2024 12:12	12.6	185.6	0.004	0	388.954	AC EF
5/27/2024 12:12	13.4	273.2	0.001	0	97.191	AC EF
5/27/2024 12:12	12.6	278	0.005	0	476.125	AC EF
5/27/2024 12:12	12.2	99.2	0.001	0	75.992	Power Line
5/27/2024 12:12	12.1	214.4	0.002	0	201.722	AC EF
5/27/2024 12:12	12.9	584.5	0.002	0	201.722	AC EF
5/27/2024 12:12	13.8	1007	0.01	0	976.08	AC EF
5/27/2024 12:12	12	844	0.009	0	851.854	Mixed
5/27/2024 12:12	12.1	874	0.005	0	480.662	AC EF
5/27/2024 12:12	12.6	922.5	0.005	0	527.008	Mixed
5/27/2024 12:12	12.3	356	0.005	0	527.008	AC EF
5/27/2024 12:12	12.9	192	0.001	0	61.196	AC EF
5/27/2024 12:12	13.4	164.2	0.005	0	471.605	AC EF
5/27/2024 12:12	12.7	73.5	0.004	0	440.449	Power Line
5/27/2024 12:12	12.5	86.1	0.003	0	316.039	Power Line
5/27/2024 12:12	13.1	298.4	0.001	0	70.039	AC EF
5/27/2024 12:12	12.6	278.6	0.002	0	212.198	AC EF
5/27/2024 12:12	11.8	289.4	0.005	0	512.917	AC EF
5/27/2024 12:12	13.1	402.8	0.001	0	75.992	AC EF
5/27/2024 12:13	12.8	523.5	0.001	0	70.039	Mixed
5/27/2024 12:13	13.1	549.5	0.001	0	64.132	AC EF
5/27/2024 12:13	12.5	189	16.558	0.002	1655768.625	AC EF
5/27/2024 12:13	11.7	455.6	2.4	0	239967.313	AC EF
5/27/2024 12:13	11.7	982	0.001	0	75.992	AC EF
5/27/2024 12:13	12	789	0.001	0	122.129	AC EF
5/27/2024 12:13	12	413	0.003	0	300.51	AC EF
5/27/2024 12:13	11.3	519	1.078	0	107761.156	Mixed
5/27/2024 12:13	12.3	1000	1.078	0	107761.156	AC EF
5/27/2024 12:13	13	281.6	0.002	0	184.528	AC EF
5/27/2024 12:13	12.4	986.5	0.005	0	536.494	AC EF
5/27/2024 12:13	12	844.5	0.005	0	462.618	AC EF
5/27/2024 12:13	12.9	268.4	0.001	0	97.191	AC EF
5/27/2024 12:13	14.7	327.2	0.001	0	73.01	Mixed
5/27/2024 12:13	13.8	327.2	5.607	0.001	560733.625	AC EF
5/27/2024 12:13	14.2	730	3.489	0	348938.563	AC EF
5/27/2024 12:13	15.2	553	0.001	0	67.08	AC EF
5/27/2024 12:13	14.6	859.5	0.001	0	147.847	AC EF
5/27/2024 12:13	13.4	554	0.001	0	125.301	Mixed
5/27/2024 12:13	14.2	519.5	0.001	0	125.301	AC EF
5/27/2024 12:13	13.4	375.2	0.001	0	144.589	AC EF
5/27/2024 12:13	13.9	526.5	0.001	0	67.08	Mixed
5/27/2024 12:13	13.1	216.8	0.005	0	517.596	AC EF
5/27/2024 12:13	13.5	178	0.009	0	869.884	AC EF
5/27/2024 12:13	13	690.5	0.005	0	480.662	AC EF
5/27/2024 12:13	12.9	125.2	0.004	0	351.865	AC EF
5/27/2024 12:13	13.5	203	0.004	0	405.856	AC EF
5/27/2024 12:13	13.2	73.2	0.109	0	10880.653	Power Line
5/27/2024 12:13	13.7	985	0.026	0	2587.969	AC EF
5/27/2024 12:13	15	966.5	0.005	0	503.612	AC EF
5/27/2024 12:13	13.7	1004.5	0.003	0	259.02	AC EF
5/27/2024 12:13	14.7	1011.5	0.006	0	550.859	AC EF
5/27/2024 12:13	15.1	1010	0.001	0	64.132	AC EF
5/27/2024 12:13	14.7	998	0.005	0	508.255	AC EF
5/27/2024 12:13	14.3	984	0.003	0	339.785	Mixed
5/27/2024 12:13	14.1	995.5	0.01	0	976.08	AC EF
5/27/2024 12:13	14.9	977.5	0.005	0	508.255	AC EF
5/27/2024 12:13	13.2	681.5	0.006	0	555.685	AC EF
5/27/2024 12:13	14.1	876.5	0.003	0	273.904	AC EF
5/27/2024 12:13	14.5	113.8	0.003	0	273.904	AC EF
5/27/2024 12:13	14.3	580	0.001	0	106.453	AC EF
5/27/2024 12:13	12.5	993	0.001	0	97.191	AC EF
5/27/2024 12:13	15.3	998.5	0.005	0	508.255	AC EF
5/27/2024 12:13	14.9	989.5	0.002	0	194.804	AC EF
5/27/2024 12:13	14.2	988	0.001	0	61.196	AC EF
5/27/2024 12:13	15	1004.5	0.01	0	1015.252	AC EF
5/27/2024 12:13	15.9	1016	0.001	0	67.08	AC EF
5/27/2024 12:13	15.7	1002	0.005	0	485.217	AC EF
5/27/2024 12:13	15.7	980.5	0.001	0	106.453	AC EF
5/27/2024 12:13	15.5	989	0.005	0	471.605	AC EF
5/27/2024 12:13	14	1001.5	0.006	0	550.859	AC EF
5/27/2024 12:13	13.4	989.5	0.001	0	67.08	Mixed
5/27/2024 12:13	12.2	985	0.011	0	1117.233	Mixed
5/27/2024 12:13	11.8	437.6	0.001	0	106.453	Mixed
5/27/2024 12:13	12.4	530	0.005	0	458.15	AC EF
5/27/2024 12:13	9.9	983.5	0.001	0	75.992	AC EF
5/27/2024 12:13	10	494.6	0.001	0	61.196	AC EF
5/27/2024 12:13	7.9	585.5	0.001	0	61.196	AC EF
5/27/2024 12:13	7.8	737.5	0.004	0	380.601	AC EF
5/27/2024 12:13	7.6	156.2	0.006	0	555.685	Mixed
5/27/2024 12:14	6.8	712.5	0.001	0	122.129	AC EF
5/27/2024 12:14	6.4	102.6	0.002	0	184.528	Mixed
5/27/2024 12:14	7	163	0.001	0	134.889	AC EF
5/27/2024 12:14	6.3	167.8	0.004	0	376.448	AC EF
5/27/2024 12:14	5.8	128	0.001	0	70.039	Mixed
5/27/2024 12:14	6.2	128	0.005	0	508.255	Mixed
5/27/2024 12:14	5.3	555.5	0.002	0	167.662	AC EF
5/27/2024 12:14	4.5	175.6	0.015	0	1469.359	Mixed
5/27/2024 12:14	4.8	121.4	0.01	0	956.831	Mixed
5/27/2024 12:14	4.6	123.6	0.008	0	822.264	Mixed
5/27/2024 12:14	4.1	92.6	0.012	0	1188.568	Mixed
5/27/2024 12:14	3.5	155.2	0.012	0	1152.557	Mixed
5/27/2024 12:14	4.5	209.6	0.003	0	304.37	Mixed
5/27/2024 12:14	3.8	489.8	0.005	0	458.15	AC EF
5/27/2024 12:14	3.8	177.8	0.006	0	565.392	Mixed
5/27/2024 12:14	4.2	184.6	0.015	0	1486.129	Mixed
5/27/2024 12:14	4.2	145.4	0.005	0	527.008	AC EF
5/27/2024 12:14	3.5	67.8	0.009	0	863.851	Mixed
5/27/2024 12:14	3.8	192.8	0.003	0	266.433	AC EF
5/27/2024 12:14	3.6	243.8	0.006	0	620.137	Mixed
5/27/2024 12:14	3	63.1	0.007	0	671.959	Power Line
5/27/2024 12:14	3.1	85.7	0.005	0	471.605	Mixed
5/27/2024 12:14	3.2	101	0.016	0	1634.011	AC EF
5/27/2024 12:14	3.5	91.2	0.003	0	273.904	Power Line
5/27/2024 12:14	3.5	91.5	0.002	0	157.697	Power Line
5/27/2024 12:14	3.6	89.2	0.011	0	1075.733	Mixed
5/27/2024 12:14	3.4	93.1	0.011	0	1055.339	Power Line
5/27/2024 12:14	3.1	90.7	0.001	0	109.564	Power Line
5/27/2024 12:14	2.7	95.9	0.001	0	106.453	Mixed
5/27/2024 12:14	2.8	79.1	0.001	0	73.01	Mixed
5/27/2024 12:14	3	78	0.001	0	73.01	Mixed
5/27/2024 12:14	2.7	75.7	0.005	0	531.742	Mixed
5/27/2024 12:14	2.6	78.3	0.013	0	1324.1	Mixed
5/27/2024 12:14	2.8	85.5	0.011	0	1145.438	Power Line
5/27/2024 12:14	2.9	75.4	0.001	0	70.039	Mixed
5/27/2024 12:14	3.1	81.7	0.013	0	1300.854	Power Line
5/27/2024 12:14	2.8	64.5	0.002	0	184.528	Power Line
5/27/2024 12:14	2.7	114.4	0.001	0	75.992	Mixed
5/27/2024 12:14	2.5	96.5	0.007	0	709.439	Mixed
5/27/2024 12:14	2.3	90.1	0.003	0	270.162	Mixed
5/27/2024 12:14	2.1	153.4	0.002	0	154.401	Mixed
5/27/2024 12:14	2.3	111.8	0.01	0	1008.66	Mixed
5/27/2024 12:14	2.5	292.4	0.004	0	423.019	Mixed
5/27/2024 12:14	2.4	237.2	0.003	0	339.785	Mixed
5/27/2024 12:14	2.2	317.6	0.011	0	1110.25	Mixed
5/27/2024 12:14	2.5	163.8	0.001	0	64.132	Mixed
5/27/2024 12:14	2.1	74.4	0.01	0	963.223	Mixed
5/27/2024 12:14	2	73.6	0.005	0	462.618	Mixed
5/27/2024 12:14	1.8	120.2	0.006	0	640.627	Mixed
5/27/2024 12:14	1.9	96.9	0.013	0	1308.573	Mixed
5/27/2024 12:14	2	73.4	0.003	0	308.245	Mixed
5/27/2024 12:14	2.1	75.3	0.001	0	144.589	Mixed
5/27/2024 12:14	1.9	73.1	0.011	0	1103.293	Mixed
5/27/2024 12:14	1.9	66.1	0.001	0	81.991	Mixed
5/27/2024 12:14	1.6	64.5	0.007	0	698.627	Mixed
5/27/2024 12:14	1.7	69.2	0.006	0	630.343	Mixed
5/27/2024 12:14	1.8	53.7	0.007	0	704.023	Mixed
5/27/2024 12:14	1.7	56.6	0.013	0	1277.875	Mixed
5/27/2024 12:14	1.8	69.9	0.064	0	6374.456	Mixed
5/27/2024 12:15	1.7	51.9	0.011	0	1068.909	Mixed
5/27/2024 12:15	1.9	57.5	0.105	0	10529.374	Mixed
5/27/2024 12:15	1.8	65.6	0.39	0	39003.902	Mixed
5/27/2024 12:15	1.6	22	3.156	0	315631.5	WiFi/Phone
5/27/2024 12:15	1.6	57.2	2.822	0	282210.438	Mixed
5/27/2024 12:15	1.6	61.1	1.545	0	154462.828	Mixed
5/27/2024 12:15	1.6	60.6	0.068	0	6795.073	Mixed
5/27/2024 12:15	1.5	75.2	0.122	0	12248.71	Mixed
5/27/2024 12:15	1.4	80.4	0.141	0	14059.814	Mixed
5/27/2024 12:15	1.6	76.2	0.056	0	5554.361	Mixed
5/27/2024 12:15	1.4	76.2	0.028	0	2796.896	Mixed
5/27/2024 12:15	1.6	74.8	0.021	0	2086.272	Mixed
5/27/2024 12:15	1.6	64.8	0.006	0	610.01	Mixed
5/27/2024 12:15	1.5	60.9	0.002	0	237.118	Mixed
5/27/2024 12:15	1.6	71.4	0.008	0	759.137	Mixed
5/27/2024 12:15	1.5	68.4	0.013	0	1347.616	Mixed
5/27/2024 12:15	1.4	66.7	0.038	0	3756.875	Mixed
5/27/2024 12:15	1.5	66.5	0.024	0	2427.441	Mixed
5/27/2024 12:15	1.5	66.5	0.024	0	2427.441	Mixed
5/27/2024 12:15	1.3	58.8	0.012	0	1232.706	Mixed
5/27/2024 12:15	1.4	60.1	0.001	0	94.128	Mixed
5/27/2024 12:15	1.5	60	0.009	0	944.121	Mixed
5/27/2024 12:15	1.5	57.5	0.013	0	1308.573	Mixed
5/27/2024 12:15	0	1.5	0	0	0	Static
5/27/2024 12:15	1.5	72.5	0.088	0	8779.366	Mixed
5/27/2024 12:15	1.5	74.2	0.046	0	4559.38	Mixed
5/27/2024 12:15	1.4	70.5	0.027	0	2743.451	Mixed
5/27/2024 12:15	1.5	53.4	0.124	0	12449.683	Mixed
5/27/2024 12:15	1.8	0	1.789	0	178910.141	WiFi/Phone
5/27/2024 12:15	1.3	75	1.125	0	112453.578	Mixed
5/27/2024 12:15	1.7	92.7	0.042	0	4150.858	Mixed
5/27/2024 12:15	1.2	94.5	0.098	0	9777.464	Mixed
5/27/2024 12:15	1.5	90.9	0.037	0	3655.092	Mixed
5/27/2024 12:15	1.3	95.7	0.021	0	2107.811	Mixed
5/27/2024 12:15	1.6	108.4	0.007	0	650.99	Mixed
5/27/2024 12:15	1.5	97	0.002	0	187.94	Mixed
5/27/2024 12:15	1.7	97.3	0.001	0	64.132	Mixed
5/27/2024 12:15	1.7	125	0.009	0	912.772	Mixed
5/27/2024 12:15	1.5	90.5	0.003	0	335.789	Mixed
5/27/2024 12:15	1.7	88.8	0.001	0	147.847	Mixed
5/27/2024 12:15	1.6	79.6	0.006	0	565.392	Mixed
5/27/2024 12:15	1.6	150.8	0.007	0	650.99	Mixed
5/27/2024 12:15	1.9	131.8	0.001	0	67.08	Mixed
5/27/2024 12:15	1.9	76.6	0.006	0	550.859	Mixed
5/27/2024 12:15	1.8	380.6	0.004	0	351.865	Mixed
5/27/2024 12:15	2.1	309.2	0.008	0	776.089	Mixed
5/27/2024 12:15	1.8	93.3	0.001	0	144.589	Mixed
5/27/2024 12:15	1.8	163	0.003	0	292.835	Mixed
5/27/2024 12:15	2.2	416	0.002	0	184.528	Mixed
5/27/2024 12:16	2	353.6	0.004	0	380.601	Mixed
5/27/2024 12:16	2.2	203	0.001	0	85.007	Mixed
5/27/2024 12:16	2	299	0.005	0	508.255	Mixed
5/27/2024 12:16	2.2	416	0.001	0	109.564	Mixed
5/27/2024 12:16	2.5	531.5	0.002	0	219.25	Mixed
5/27/2024 12:16	2.2	582	0.006	0	610.01	Mixed
5/27/2024 12:16	2.6	365.6	0.001	0	147.847	Mixed
5/27/2024 12:16	2.9	409.4	0.001	0	64.132	Mixed
5/27/2024 12:16	3	357.8	0.011	0	1110.25	AC EF
5/27/2024 12:16	3	321.8	0.011	0	1103.293	AC EF
5/27/2024 12:16	3	335	0.006	0	604.976	AC EF
5/27/2024 12:16	2.8	302	0.005	0	498.987	Power Line
5/27/2024 12:16	3.1	101.2	0.003	0	251.663	AC EF
5/27/2024 12:16	3.2	488	0.004	0	414.404	Mixed
5/27/2024 12:16	3.1	396.2	0.005	0	541.264	Mixed
5/27/2024 12:16	3.2	545.5	0.005	0	541.264	AC EF
5/27/2024 12:16	3.6	905.5	0.001	0	115.822	Mixed
5/27/2024 12:16	3.9	970	0.002	0	171.009	Mixed
5/27/2024 12:16	4	992	0.005	0	522.293	Mixed
5/27/2024 12:16	4.8	993.5	0.002	0	174.369	Mixed
5/27/2024 12:16	4.9	1007.5	0.002	0	244.363	Mixed
5/27/2024 12:16	5.4	1009.5	0.005	0	489.789	AC EF
5/27/2024 12:16	6.4	1003	0.005	0	512.917	Mixed
5/27/2024 12:16	7.4	1000	0.001	0	81.991	AC EF
5/27/2024 12:16	9.5	996	0.005	0	527.008	AC EF
5/27/2024 12:16	10.5	993	0.003	0	335.789	AC EF
5/27/2024 12:16	10.9	990	0.001	0	100.267	AC EF
5/27/2024 12:16	12.1	1006	0.001	0	100.267	AC EF
5/27/2024 12:16	13.6	1019.5	0.001	0	58.27	AC EF
5/27/2024 12:16	13.6	1016.5	0.001	0	147.847	AC EF
5/27/2024 12:16	12.9	1012	0.002	0	154.401	AC EF
5/27/2024 12:16	13.5	1005.5	0.003	0	300.51	AC EF
5/27/2024 12:16	12.8	1005	0.002	0	154.401	AC EF
5/27/2024 12:16	12.7	994	0.006	0	625.23	AC EF
5/27/2024 12:16	12.8	996.5	0.001	0	81.991	Mixed
5/27/2024 12:16	12.9	991.5	0.002	0	171.009	AC EF
5/27/2024 12:16	12.8	1006	0.003	0	296.665	AC EF
5/27/2024 12:16	13	1020	0.004	0	397.372	Mixed
5/27/2024 12:16	13.6	1019	0.001	0	109.564	AC EF
5/27/2024 12:16	14.3	1016	0.001	0	122.129	AC EF
5/27/2024 12:16	13.5	1010.5	0.001	0	147.847	AC EF
5/27/2024 12:16	14.5	1004	0.001	0	64.132	AC EF
5/27/2024 12:16	14.1	999.5	0.001	0	144.589	AC EF
5/27/2024 12:16	13.9	985.5	0.001	0	64.132	AC EF
5/27/2024 12:16	14.3	996	0.003	0	273.904	AC EF
5/27/2024 12:16	14.1	1012.5	0.005	0	503.612	AC EF
5/27/2024 12:16	12.9	1012.5	0.005	0	498.987	AC EF
5/27/2024 12:16	12.5	1007	0.004	0	401.606	Mixed
5/27/2024 12:16	12.5	1002	0.003	0	259.02	AC EF
5/27/2024 12:16	13.2	992	0.001	0	75.992	AC EF
5/27/2024 12:16	12	982	0.001	0	64.132	AC EF
5/27/2024 12:16	12.8	980	0.001	0	70.039	AC EF
5/27/2024 12:16	12.8	975	0.004	0	359.996	AC EF
5/27/2024 12:16	12.9	1008	0.001	0	64.132	AC EF
5/27/2024 12:16	13.3	994	0.004	0	372.311	AC EF
5/27/2024 12:16	12.9	1009.5	0.001	0	147.847	AC EF
5/27/2024 12:16	12.1	988.5	0.001	0	64.132	AC EF
5/27/2024 12:16	12.5	980	0.005	0	485.217	AC EF
5/27/2024 12:16	13.2	998	0.005	0	541.264	AC EF
5/27/2024 12:16	14.5	986	0.001	0	81.991	AC EF
5/27/2024 12:17	14.5	1015	0.001	0	147.847	AC EF
5/27/2024 12:17	12.7	1020.5	0.002	0	161.006	AC EF
5/27/2024 12:17	13.1	1009	0.006	0	565.392	AC EF
5/27/2024 12:17	13.1	1004	0.005	0	476.125	AC EF
5/27/2024 12:17	13	867	0.004	0	351.865	Mixed
5/27/2024 12:17	15	999.5	0.005	0	508.255	AC EF
5/27/2024 12:17	15.9	991	0.009	0	888.123	AC EF
5/27/2024 12:17	15.6	987	0.001	0	147.847	AC EF
5/27/2024 12:17	14.9	844	0.005	0	458.15	AC EF
5/27/2024 12:17	12.1	1005.5	0.002	0	181.129	AC EF
5/27/2024 12:17	13.4	191.6	0.001	0	103.354	Mixed
5/27/2024 12:17	12.4	885.5	0.002	0	194.804	AC EF
5/27/2024 12:17	10.8	983.5	0.001	0	67.08	AC EF
5/27/2024 12:17	12	253.4	0.005	0	522.293	AC EF
5/27/2024 12:17	11.6	366.8	0.001	0	75.992	AC EF
5/27/2024 12:17	11.1	1000.5	0.001	0	64.132	Mixed
5/27/2024 12:17	10.6	1017.5	0.002	0	177.743	Mixed
5/27/2024 12:17	12.1	149.2	0.005	0	512.917	Mixed
5/27/2024 12:17	11.8	551.5	0.005	0	494.379	AC EF
5/27/2024 12:17	9.5	910.5	0.001	0	85.007	AC EF
5/27/2024 12:17	8.8	1000.5	0.01	0	976.08	AC EF
5/27/2024 12:17	8.9	985.5	0.005	0	503.612	AC EF
5/27/2024 12:17	6.7	987.5	0.006	0	610.01	Mixed
5/27/2024 12:17	6.6	175	0.011	0	1075.733	Mixed
5/27/2024 12:17	6	336.8	0.007	0	704.023	AC EF
5/27/2024 12:17	5.8	773.5	0.005	0	546.052	AC EF
5/27/2024 12:17	5.3	569	0.009	0	912.772	Power Line
5/27/2024 12:17	4.7	453.8	0.005	0	531.742	AC EF
5/27/2024 12:17	4.5	778	0.006	0	580.093	AC EF
5/27/2024 12:17	4.4	414.2	0.006	0	585.032	AC EF
5/27/2024 12:17	4.2	141.6	0.004	0	368.191	Mixed
5/27/2024 12:17	4	118.2	0.006	0	585.032	Mixed
5/27/2024 12:17	4	76.9	0.007	0	677.251	Power Line
5/27/2024 12:17	3.5	422.6	0.001	0	64.132	AC EF
5/27/2024 12:17	3.2	341	0.001	0	100.267	Mixed
5/27/2024 12:17	3	243.2	0.01	0	995.553	AC EF
5/27/2024 12:17	3.2	204.2	0.001	0	97.191	Mixed
5/27/2024 12:17	2.9	251.6	0.003	0	270.162	Mixed
5/27/2024 12:17	3	447.8	0.013	0	1300.854	Mixed
5/27/2024 12:17	2.9	366.8	0.001	0	100.267	Mixed
5/27/2024 12:17	2.7	274.4	0.003	0	296.665	Mixed
5/27/2024 12:17	2.4	153.4	0.008	0	776.089	Mixed
5/27/2024 12:17	2.6	136.4	0.001	0	100.267	Mixed
5/27/2024 12:17	2.4	134.4	0.01	0	1002.094	Mixed
5/27/2024 12:17	2.3	162.4	0.006	0	560.529	Mixed
5/27/2024 12:17	2	108	0.006	0	610.01	Mixed
5/27/2024 12:17	2.5	91.6	48.678	0.005	4867807	Mixed
5/27/2024 12:17	2.2	93.8	7.429	0.001	742949.25	Mixed
5/27/2024 12:17	2	95.2	6.057	0.001	605676.25	Mixed
5/27/2024 12:17	1.9	149.4	2.715	0	271524.438	Mixed
5/27/2024 12:17	2.1	85.6	0.003	0	316.039	Mixed
5/27/2024 12:17	1.8	103.4	0.012	0	1188.568	Mixed
5/27/2024 12:17	2	128.8	0.002	0	174.369	Mixed
5/27/2024 12:17	1.7	135.8	0.013	0	1308.573	Mixed
5/27/2024 12:17	1.9	152	0.002	0	212.198	Mixed
5/27/2024 12:17	1.8	126	0.005	0	531.742	Mixed
5/27/2024 12:17	1.9	94.1	0.001	0	131.681	Mixed
5/27/2024 12:17	1.8	65.6	0.014	0	1371.405	Mixed
5/27/2024 12:17	1.8	70	0.024	0	2356.019	Mixed
5/27/2024 12:17	1.8	69.7	0.003	0	266.433	Mixed
5/27/2024 12:18	1.8	71.9	0.005	0	508.255	Mixed
5/27/2024 12:18	1.8	65.6	0.003	0	339.785	Mixed
5/27/2024 12:18	1.6	66.4	0.005	0	489.789	Mixed
5/27/2024 12:18	1.8	71.9	0.007	0	671.959	Mixed
5/27/2024 12:18	1.9	76.6	0.004	0	364.086	Mixed
5/27/2024 12:18	1.9	93.9	0.007	0	671.959	Mixed
5/27/2024 12:18	1.8	91.5	0.005	0	480.662	Mixed
5/27/2024 12:18	1.6	84.6	0.004	0	359.996	Mixed
5/27/2024 12:18	1.4	65.4	0.004	0	397.372	Mixed
5/27/2024 12:18	1.4	66.8	0.012	0	1159.704	Mixed
5/27/2024 12:18	2.3	74.5	0.059	0	5875.389	Mixed
5/27/2024 12:18	1.4	66.4	0.035	0	3506.759	Mixed
5/27/2024 12:18	1.8	68.6	0.012	0	1152.557	Mixed
5/27/2024 12:18	1.7	63.7	0.006	0	645.798	Mixed
5/27/2024 12:18	1.6	56.8	0.009	0	944.121	Mixed
5/27/2024 12:18	2	50.1	0.036	0	3621.684	Mixed
5/27/2024 12:18	1.8	23.5	0.012	0	1210.509	Mixed
5/27/2024 12:18	1.7	66.8	0.002	0	157.697	Mixed
5/27/2024 12:18	1.6	61.9	0.011	0	1117.233	Mixed
5/27/2024 12:18	1.5	60	0.018	0	1772.781	Mixed
5/27/2024 12:18	1.6	60.2	0.05	0	4958.563	Mixed
5/27/2024 12:18	1.6	60.2	0.05	0	4958.563	Mixed
5/27/2024 12:18	1.6	60.2	0.05	0	4958.563	Mixed
5/27/2024 12:18	1.6	60.2	0.05	0	4958.563	Mixed
5/27/2024 12:18	1.6	14.5	0.026	0	2562.746	Mixed
5/27/2024 12:18	1.6	2.1	0.032	0	3180	Mixed
5/27/2024 12:18	1.5	2.5	0.002	0	212.198	Mixed
5/27/2024 12:18	1.5	1	0.005	0	531.742	Mixed
5/27/2024 12:18	1.7	7.5	0.131	0	13124.637	Mixed
5/27/2024 12:18	1.5	19	1.273	0	127290.664	WiFi/Phone
5/27/2024 12:18	1.7	5.5	0.086	0	8634.488	Mixed
5/27/2024 12:18	1.6	0.2	1.65	0	164958.938	WiFi/Phone
5/27/2024 12:18	1.6	0	3.74	0	374030.188	WiFi/Phone
5/27/2024 12:18	1.6	1.7	4.449	0	444922.531	WiFi/Phone
5/27/2024 12:18	1.6	6.5	3.306	0	330590.281	WiFi/Phone
5/27/2024 12:18	1.7	1.6	1.874	0	187403.594	WiFi/Phone
5/27/2024 12:18	1.7	3	0.065	0	6456.646	Mixed
5/27/2024 12:18	1.7	1.2	0.044	0	4380.247	Mixed
5/27/2024 12:18	1.7	0.6	0.028	0	2796.896	Mixed
5/27/2024 12:18	1.7	12.5	0.045	0	4519.034	Mixed
5/27/2024 12:18	1.4	50.5	0.139	0	13946.59	Mixed
5/27/2024 12:18	1.5	1	0.08	0	8009.458	Mixed
5/27/2024 12:18	1.4	0	0.205	0	20496.617	Mixed
5/27/2024 12:18	1.7	0	0.362	0	36202.996	Mixed
5/27/2024 12:18	1.1	50.1	0.015	0	1486.129	Mixed
5/27/2024 12:18	1.7	0	0.008	0	828.137	Mixed
5/27/2024 12:18	1.4	0	0.013	0	1300.854	Mixed
5/27/2024 12:18	1.6	0	0.022	0	2151.39	Mixed
5/27/2024 12:18	1.7	0	0.046	0	4600.037	Mixed
5/27/2024 12:18	1.3	0	0.017	0	1707.024	Mixed
5/27/2024 12:18	1.7	0	0.026	0	2639	Mixed
5/27/2024 12:18	1.5	2.7	0.04	0	4021.862	Mixed
5/27/2024 12:18	1.5	0	0.017	0	1670.236	Mixed
5/27/2024 12:18	1.4	0	0.071	0	7149.501	Mixed
5/27/2024 12:18	1.7	0	0.031	0	3091.391	Mixed
5/27/2024 12:18	1.4	0	0.005	0	541.264	Mixed
5/27/2024 12:18	1.5	2.5	0.013	0	1277.875	Mixed
5/27/2024 12:19	1.6	0	0.017	0	1679.38	Mixed
5/27/2024 12:19	1.2	8.5	0.08	0	8009.458	Mixed
5/27/2024 12:19	1.8	2.1	0.004	0	414.404	Mixed
5/27/2024 12:19	1.6	6.5	0.021	0	2118.643	Mixed
5/27/2024 12:19	1.4	16.5	0.002	0	174.369	Mixed
5/27/2024 12:19	1.6	52.5	0.013	0	1347.616	Mixed
5/27/2024 12:19	1.6	50.5	0.007	0	731.314	Mixed
5/27/2024 12:19	1.5	60.4	0.015	0	1494.562	Mixed
5/27/2024 12:19	1.5	11	0.03	0	3033.445	Mixed
5/27/2024 12:19	1.4	1.9	0.036	0	3555.634	Mixed
5/27/2024 12:19	1.2	0.4	0.01	0	989.037	Mixed
5/27/2024 12:19	1.5	1.3	0.009	0	925.239	Mixed
5/27/2024 12:19	1.7	0.6	0.003	0	327.843	Mixed
5/27/2024 12:19	1.7	56.6	0.013	0	1285.505	Mixed
5/27/2024 12:19	1.6	52.6	0.007	0	682.564	Mixed
5/27/2024 12:19	10	78.8	1.082	0	108179.57	Mixed
5/27/2024 12:19	6	64	0.001	0	64.132	Power Line
5/27/2024 12:19	3.5	89	0.001	0	64.132	Power Line
5/27/2024 12:19	2.1	77.2	0.04	0	4021.862	Mixed
5/27/2024 12:19	5	78.3	0.046	0	4579.669	Mixed
5/27/2024 12:19	9.8	78.5	0.006	0	555.685	Mixed
5/27/2024 12:19	1.1	9.5	0.01	0	1035.18	Mixed
5/27/2024 12:19	1.7	4.5	0.003	0	277.661	Mixed
5/27/2024 12:19	1.7	2.1	1.682	0	168179.063	Mixed
5/27/2024 12:19	1.4	3.5	1.631	0	163056.422	WiFi/Phone
5/27/2024 12:19	1.5	0	1.708	0	170800.172	WiFi/Phone
5/27/2024 12:19	1.5	1.3	1.328	0	132828	Mixed
5/27/2024 12:19	1.3	0.8	1.133	0	113328.297	Mixed
5/27/2024 12:19	1.7	0	1.925	0	192542.375	WiFi/Phone
5/27/2024 12:19	1.4	0	0.919	0	91922.078	Mixed
5/27/2024 12:19	1.3	1.3	3.887	0	388740.031	WiFi/Phone
5/27/2024 12:19	1.5	33.3	3.37	0	337030	Mixed
5/27/2024 12:19	1.6	90.8	2.899	0	289939.281	Mixed
5/27/2024 12:19	1.7	151.2	0.007	0	677.251	Mixed
5/27/2024 12:19	1.7	190	0.015	0	1528.623	Mixed
5/27/2024 12:19	1.4	77.5	0.013	0	1270.274	Mixed
5/27/2024 12:19	1.1	27.5	0.006	0	645.798	Mixed
5/27/2024 12:19	1.6	1	0.006	0	589.989	Mixed
5/27/2024 12:19	1.3	0	0.011	0	1110.25	Mixed
5/27/2024 12:19	1.5	0.8	0.026	0	2626.168	Mixed
5/27/2024 12:19	1.6	0.4	0.009	0	875.94	Mixed
5/27/2024 12:19	1.4	0.6	0.012	0	1181.31	Mixed
5/27/2024 12:19	1.5	0.2	0.011	0	1075.733	Mixed
5/27/2024 12:19	1.6	0	0.026	0	2639	Mixed
5/27/2024 12:19	1.5	0.2	0.005	0	489.789	Mixed
5/27/2024 12:19	1.5	0.2	0.004	0	401.606	Mixed
5/27/2024 12:19	1.7	0.4	0.011	0	1062.111	Mixed
5/27/2024 12:19	1.4	0	0.013	0	1339.747	Mixed
5/27/2024 12:19	1.7	1.5	0.018	0	1782.321	Mixed
5/27/2024 12:19	1.6	0.6	0.012	0	1152.557	Mixed
5/27/2024 12:19	1.5	2	0.025	0	2488.233	Mixed
5/27/2024 12:19	1.5	0	0.045	0	4498.978	Mixed
5/27/2024 12:19	1.5	5.5	0.007	0	731.314	Mixed
5/27/2024 12:19	1.5	5	0.063	0	6293.208	Mixed
5/27/2024 12:19	1.6	0.2	0.015	0	1486.129	Mixed
5/27/2024 12:19	1.8	1.7	0.008	0	798.999	Mixed
5/27/2024 12:19	1.8	21.5	0.007	0	687.898	Mixed
5/27/2024 12:19	1.8	2.1	0.047	0	4744.829	Mixed
5/27/2024 12:19	1.8	0	3.449	0	344923.094	WiFi/Phone
5/27/2024 12:19	1.4	0	6.246	0.001	624645	Cell Tower
5/27/2024 12:20	1.1	1.3	6.932	0.001	693155.063	Cell Tower
5/27/2024 12:20	1.3	0	2.336	0	233567.359	Cell Tower
5/27/2024 12:20	1.6	2.6	0.222	0	22193.4	Mixed
5/27/2024 12:20	1.7	22	1.138	0	113768.195	WiFi/Phone
5/27/2024 12:20	1.3	1.1	5.987	0.001	598712.25	WiFi/Phone
5/27/2024 12:20	1.5	0	17.076	0.002	1707586.5	WiFi/Phone
5/27/2024 12:20	1.6	0	1.545	0	154462.828	WiFi/Phone
5/27/2024 12:20	1.5	0	2.327	0	232667.094	WiFi/Phone
5/27/2024 12:20	1.5	0	3.798	0	379846.25	Cell Tower
5/27/2024 12:20	1.5	0	4.232	0	423166.125	WiFi/Phone
5/27/2024 12:20	1.4	0	3.902	0	390242.375	Cell Tower
5/27/2024 12:20	1.1	51.8	7.176	0.001	717620.875	Mixed
5/27/2024 12:20	1.8	0.6	28.832	0.003	2883187.25	Cell Tower
5/27/2024 12:20	1.4	1.3	7.963	0.001	796316.875	WiFi/Phone
5/27/2024 12:20	1.8	23	1.669	0	166883.578	WiFi/Phone
5/27/2024 12:20	1.4	0	8.118	0.001	811809.5	WiFi/Phone
5/27/2024 12:20	1.6	0	5.395	0.001	539527.375	WiFi/Phone
5/27/2024 12:20	1.8	0	3.858	0	385752.563	WiFi/Phone
5/27/2024 12:20	1.5	0	8.437	0.001	843704.125	WiFi/Phone
5/27/2024 12:20	1.6	0	0.002	0	194.804	Mixed
5/27/2024 12:20	1.3	0	1.065	0	106515.516	Mixed
5/27/2024 12:20	1.6	1.8	0.016	0	1554.515	Mixed
5/27/2024 12:20	2	0	1.557	0	155662.313	WiFi/Phone
5/27/2024 12:20	3	0	0.002	0	215.717	Mixed
5/27/2024 12:20	1.6	1.9	0.009	0	918.993	Mixed
5/27/2024 12:20	1.4	0	0.009	0	851.854	Mixed
5/27/2024 12:20	1.3	0	0.082	0	8247.392	Mixed
5/27/2024 12:20	1.4	0	0.002	0	161.006	Mixed
5/27/2024 12:20	1.5	0	0.017	0	1697.774	Mixed
5/27/2024 12:20	1.4	0.2	0.001	0	78.985	Mixed
5/27/2024 12:20	1.5	0	0.004	0	410.122	Mixed
5/27/2024 12:20	1.4	0	0.035	0	3458.446	Mixed
5/27/2024 12:20	1.7	0	0.001	0	112.687	Mixed
5/27/2024 12:20	1.4	0	0.001	0	70.039	Mixed
5/27/2024 12:20	1.1	0	0.015	0	1494.562	Mixed
5/27/2024 12:20	1.6	1	0.029	0	2864.869	Mixed
5/27/2024 12:20	1.4	0	0.004	0	444.849	Mixed
5/27/2024 12:20	1.7	0	0.001	0	70.039	Mixed
5/27/2024 12:20	1.4	0	0.011	0	1096.363	Mixed
5/27/2024 12:20	1.6	0	0.013	0	1277.875	Mixed
5/27/2024 12:20	1.4	0	0.003	0	308.245	Mixed
5/27/2024 12:20	1.2	0	0.016	0	1589.507	Mixed
5/27/2024 12:20	1.6	0	0.005	0	517.596	Mixed
5/27/2024 12:20	1.6	0	0.005	0	453.699	Mixed
5/27/2024 12:20	1.1	1.3	0.001	0	94.128	Mixed
5/27/2024 12:20	1.4	0.6	0.023	0	2286.228	Mixed
5/27/2024 12:20	1.6	0	0.001	0	112.687	Mixed
5/27/2024 12:20	1.6	0	0.016	0	1598.339	Mixed
5/27/2024 12:20	1.3	0	0.002	0	237.118	Mixed
5/27/2024 12:20	1.2	0	0.007	0	677.251	Mixed
5/27/2024 12:20	1.6	0	0.007	0	677.251	Mixed
5/27/2024 12:20	1.7	1	0.008	0	839.95	Mixed
5/27/2024 12:20	1.6	0	0.008	0	793.238	Mixed
5/27/2024 12:20	1.5	1	0.005	0	522.293	Mixed
5/27/2024 12:20	1.2	0	0.029	0	2892.427	Mixed
5/27/2024 12:20	1.5	2.4	0.001	0	106.453	Mixed
5/27/2024 12:20	1.4	1.4	0.006	0	575.174	Mixed
5/27/2024 12:20	1.4	2.5	0.003	0	343.796	Mixed
5/27/2024 12:20	1.3	0	0.009	0	900.4	Mixed
5/27/2024 12:20	1.5	0	0.005	0	471.605	Mixed
5/27/2024 12:21	1.4	0	0.006	0	550.859	Mixed
5/27/2024 12:21	1.3	27	0.009	0	888.123	Mixed
5/27/2024 12:21	1.2	0	0.005	0	498.987	Mixed
5/27/2024 12:21	1.5	0	0.001	0	64.132	Mixed
5/27/2024 12:21	1.4	0	0.106	0	10572.694	Mixed
5/27/2024 12:21	1.8	0.6	0.003	0	304.37	Mixed
5/27/2024 12:21	1.4	53.2	0.015	0	1503.028	Mixed
5/27/2024 12:21	1.5	10.5	0.806	0	80556.875	Mixed
5/27/2024 12:21	1.6	1.6	0.02	0	1950.239	Mixed
5/27/2024 12:21	1.6	2.5	0.073	0	7302.083	Mixed
5/27/2024 12:21	1.5	0.6	0.015	0	1528.623	Mixed
5/27/2024 12:21	1.6	1.5	0.009	0	869.884	Mixed
5/27/2024 12:21	1.3	1.4	0.015	0	1511.527	Mixed
5/27/2024 12:21	1.3	1.1	0.009	0	937.803	Mixed
5/27/2024 12:21	1.7	1.5	0.016	0	1563.213	Mixed
5/27/2024 12:21	1.5	1.9	0.025	0	2451.618	Mixed
5/27/2024 12:21	1.6	0.2	0.018	0	1820.848	Mixed
5/27/2024 12:21	1.6	1.5	0.021	0	2129.517	Mixed
5/27/2024 12:21	1.4	0	0.056	0	5578.49	Mixed
5/27/2024 12:21	1.4	2.6	0.003	0	285.219	Mixed
5/27/2024 12:21	1.4	0.8	0.004	0	449.265	Mixed
5/27/2024 12:21	1.3	0.8	0.002	0	187.94	Mixed
5/27/2024 12:21	1.5	0	0.001	0	112.687	Mixed
5/27/2024 12:21	1.3	0	0.014	0	1395.471	Mixed
5/27/2024 12:21	1.3	0	0.001	0	81.991	Mixed
5/27/2024 12:21	1.2	0	0.001	0	67.08	Mixed
5/27/2024 12:21	1.4	1.4	0.018	0	1791.897	Mixed
5/27/2024 12:21	1.3	0	0.011	0	1103.293	Mixed
5/27/2024 12:21	1.5	0	0.013	0	1308.573	Mixed
5/27/2024 12:21	1.3	0	0.001	0	61.196	Mixed
5/27/2024 12:21	1.6	0	0.015	0	1494.562	Mixed
5/27/2024 12:21	1.1	2.4	0.012	0	1188.568	Mixed
5/27/2024 12:21	1.4	2.8	0.019	0	1869.849	Mixed
5/27/2024 12:21	1.7	0	0.002	0	151.118	Mixed
5/27/2024 12:21	1.2	2.7	0.007	0	677.251	Mixed
5/27/2024 12:21	1.3	29.8	0.012	0	1152.557	Mixed
5/27/2024 12:21	1.1	54.2	0.005	0	498.987	Mixed
5/27/2024 12:21	1.7	2.6	0.026	0	2600.652	Mixed
5/27/2024 12:21	1.3	0	0.003	0	300.51	Mixed
5/27/2024 12:21	1.2	2.3	0.011	0	1068.909	Mixed
5/27/2024 12:21	1.6	24	0.004	0	449.265	Mixed
5/27/2024 12:21	1.2	0	0.005	0	527.008	Mixed
5/27/2024 12:21	1.4	0	0.147	0	14698.378	Mixed
5/27/2024 12:21	1.5	0	0.004	0	393.155	Mixed
5/27/2024 12:21	1.3	0.8	0.05	0	5002.306	Mixed
5/27/2024 12:21	1.5	1	0.005	0	503.612	Mixed
5/27/2024 12:21	1.4	0	0.025	0	2475.981	Mixed
5/27/2024 12:21	1.4	0	0.013	0	1324.1	Mixed
5/27/2024 12:21	1.3	0	0.008	0	764.766	Mixed
5/27/2024 12:21	1.1	0	0.006	0	575.174	Mixed
5/27/2024 12:21	1.4	0	0.015	0	1486.129	Mixed
5/27/2024 12:21	1.4	1	0.021	0	2129.517	Mixed
5/27/2024 12:21	1.5	0	0.022	0	2240.588	Mixed
5/27/2024 12:21	1.1	0	0.006	0	575.174	Mixed
5/27/2024 12:21	1.4	0	0.004	0	401.606	Mixed
5/27/2024 12:21	1.3	0	0.01	0	982.546	Mixed
5/27/2024 12:21	1.5	0	0.042	0	4207.216	Mixed
5/27/2024 12:21	1.4	7.5	0.002	0	177.743	Mixed
5/27/2024 12:21	1.3	0	0.001	0	64.132	Mixed
5/27/2024 12:21	1.5	0	0.004	0	418.704	Mixed
5/27/2024 12:22	1.3	0	0.003	0	319.959	Mixed
5/27/2024 12:22	1.6	0	0.016	0	1580.708	Mixed
5/27/2024 12:22	1.5	0.6	0.011	0	1068.909	Mixed
5/27/2024 12:22	1	1.6	0.005	0	503.612	Mixed
5/27/2024 12:22	1.6	1.1	0.002	0	177.743	Mixed
5/27/2024 12:22	1.2	0.2	0.029	0	2892.427	Mixed
5/27/2024 12:22	1.1	1.5	0.017	0	1716.31	Mixed
5/27/2024 12:22	1.5	1.3	0.014	0	1444.444	Mixed
5/27/2024 12:22	1	0	0.002	0	171.009	Mixed
5/27/2024 12:22	1.3	1.1	0.014	0	1444.444	Mixed
5/27/2024 12:22	1.5	1.6	0.022	0	2206.816	Mixed
5/27/2024 12:22	1.3	0	0.008	0	776.089	Mixed
5/27/2024 12:22	1.4	67.2	0.007	0	720.335	Mixed
5/27/2024 12:22	1.4	0	0.006	0	585.032	Mixed
5/27/2024 12:22	1.6	0	0.001	0	61.196	Mixed
5/27/2024 12:22	1.3	71.2	0.024	0	2439.506	Mixed
5/27/2024 12:22	1.5	0	0.049	0	4893.576	Mixed
5/27/2024 12:22	1.3	63.1	0.001	0	85.007	Mixed
5/27/2024 12:22	0.9	0	0.001	0	73.01	Mixed
5/27/2024 12:22	1.3	0	0.024	0	2427.441	Mixed
5/27/2024 12:22	1.2	62.2	0.002	0	222.796	Mixed
5/27/2024 12:22	1.3	0	0.003	0	270.162	Mixed
5/27/2024 12:22	1.4	0	0.009	0	894.25	Mixed
5/27/2024 12:22	1.1	2.1	0.001	0	122.129	Mixed
5/27/2024 12:22	1.2	0.8	0.08	0	7975.989	Mixed
5/27/2024 12:22	1.6	0	0.08	0	7975.989	Mixed
5/27/2024 12:22	1.4	0	0.009	0	906.574	Mixed
5/27/2024 12:22	1.3	0	0.008	0	839.95	Mixed
5/27/2024 12:22	1.3	0	0.006	0	580.093	Mixed
5/27/2024 12:22	1.5	0	0.057	0	5651.434	Mixed
5/27/2024 12:22	1.4	0	0.004	0	418.704	Mixed
5/27/2024 12:22	1.3	1.4	0.006	0	599.961	Mixed
5/27/2024 12:22	1.4	52.2	0.006	0	599.961	Mixed
5/27/2024 12:22	1.4	0	0.004	0	380.601	Mixed
5/27/2024 12:22	1.5	0.6	0.009	0	869.884	Mixed
5/27/2024 12:22	1.3	0	0.001	0	64.132	Mixed
5/27/2024 12:22	1.2	1.2	0.002	0	215.717	Mixed
5/27/2024 12:22	1.3	0.8	0.001	0	85.007	Mixed
5/27/2024 12:22	0.9	0	0.006	0	560.529	Mixed
5/27/2024 12:22	1.5	0	0.006	0	635.475	Mixed
5/27/2024 12:22	1.5	0	0.011	0	1096.363	Mixed
5/27/2024 12:22	1.5	0	0.003	0	262.719	Mixed
5/27/2024 12:22	1.4	0	0.007	0	661.434	Mixed
5/27/2024 12:22	1.2	0	0.003	0	259.02	Mixed
5/27/2024 12:22	1.5	0	0.008	0	834.032	Mixed
5/27/2024 12:22	1.2	0	0.012	0	1225.278	Mixed
5/27/2024 12:22	1.4	0.8	0.009	0	918.993	Mixed
5/27/2024 12:22	1.5	1	0.009	0	894.25	Mixed
5/27/2024 12:22	0.9	0	0.005	0	517.596	Mixed
5/27/2024 12:22	1.4	0.6	0.004	0	397.372	Mixed
5/27/2024 12:22	1	1.1	0.003	0	308.245	Mixed
5/27/2024 12:22	1.1	1	0.001	0	75.992	Mixed
5/27/2024 12:22	1.3	0	0.012	0	1203.167	Mixed
5/27/2024 12:22	1.2	0.6	0.023	0	2251.932	Mixed
5/27/2024 12:22	1.3	0	0.002	0	248.006	Mixed
5/27/2024 12:22	1.1	0	0.004	0	384.77	Mixed
5/27/2024 12:22	1.2	0	0.008	0	828.137	Mixed
5/27/2024 12:22	1.1	0	0.017	0	1716.31	Mixed
5/27/2024 12:22	1.4	0	0.02	0	1950.239	Mixed
5/27/2024 12:22	1.3	0	0.016	0	1554.515	Mixed
5/27/2024 12:23	1.3	0	0.003	0	327.843	Mixed
5/27/2024 12:23	1.5	0.6	0.016	0	1625.042	Mixed
5/27/2024 12:23	1.1	0	0.022	0	2229.287	Mixed
5/27/2024 12:23	1.4	0	0.018	0	1782.321	Mixed
5/27/2024 12:23	1.4	0	0.033	0	3285.998	Mixed
5/27/2024 12:23	1.2	0.2	0.038	0	3843.509	Mixed
5/27/2024 12:23	1.4	0	0.028	0	2837.522	Mixed
5/27/2024 12:23	1.6	15	0.015	0	1461.022	Mixed
5/27/2024 12:23	1.3	0	0.009	0	894.25	Mixed
5/27/2024 12:23	1.5	0	0.006	0	599.961	Mixed
5/27/2024 12:23	1.4	0	0.016	0	1571.944	Mixed
5/27/2024 12:23	1.3	1.7	0.005	0	476.125	Mixed
5/27/2024 12:23	1.5	2.7	0.008	0	776.089	Mixed
5/27/2024 12:23	1.3	2.5	0.001	0	94.128	Mixed
5/27/2024 12:23	1.5	5	0.009	0	906.574	Mixed
5/27/2024 12:23	1.5	0	0.001	0	131.681	Mixed
5/27/2024 12:23	1.2	1.5	0.006	0	560.529	Mixed
5/27/2024 12:23	1.4	18.5	0.02	0	1950.239	Mixed
5/27/2024 12:23	1.4	0	0.01	0	963.223	Mixed
5/27/2024 12:23	1.5	0	0.014	0	1403.555	Mixed
5/27/2024 12:23	1.5	0	0.009	0	937.803	Mixed
5/27/2024 12:23	1.3	0	0.009	0	918.993	Mixed
5/27/2024 12:23	1.4	0.2	0.008	0	804.782	Mixed
5/27/2024 12:23	1.6	0	0.001	0	128.484	Mixed
5/27/2024 12:23	1.2	0.8	0.012	0	1152.557	Mixed
5/27/2024 12:23	1.3	0.6	0.002	0	198.256	Mixed
5/27/2024 12:23	1.1	1	0.014	0	1427.994	Mixed
5/27/2024 12:23	1.5	0	0.003	0	266.433	Mixed
5/27/2024 12:23	1.3	1	0.006	0	560.529	Mixed
5/27/2024 12:23	1.3	0	0.005	0	458.15	Mixed
5/27/2024 12:23	1	0	0.003	0	285.219	Mixed
5/27/2024 12:23	1.1	0	0.026	0	2562.746	Mixed
5/27/2024 12:23	1.4	0	0.003	0	300.51	Mixed
5/27/2024 12:23	1.1	27.5	0.022	0	2184.518	Mixed
5/27/2024 12:23	1.7	0	0.007	0	704.023	Mixed
5/27/2024 12:23	1.2	0.6	0.002	0	222.796	Mixed
5/27/2024 12:23	1.1	0	0.004	0	449.265	Mixed
5/27/2024 12:23	1.2	0	0.005	0	485.217	Mixed
5/27/2024 12:23	1.2	0.6	0.018	0	1772.781	Mixed
5/27/2024 12:23	1.1	0	0.016	0	1598.339	Mixed
5/27/2024 12:23	1.3	15.5	0.002	0	177.743	Mixed
5/27/2024 12:23	1.1	1.5	0.018	0	1791.897	Mixed
5/27/2024 12:23	1.4	50.2	0.002	0	187.94	Mixed
5/27/2024 12:23	1.2	0	0.019	0	1889.715	Mixed
5/27/2024 12:23	1.2	0	0.013	0	1262.702	Mixed
5/27/2024 12:23	1.6	62	0.013	0	1262.702	Mixed
5/27/2024 12:23	0.8	0	0.02	0	2043.688	Mixed
5/27/2024 12:23	1.3	0	0.111	0	11060.392	Mixed
5/27/2024 12:23	1.1	0	0.005	0	522.293	Mixed
5/27/2024 12:23	1.2	2.1	0.015	0	1503.028	Mixed
5/27/2024 12:23	1.1	0	0.99	0	98952.016	Mixed
5/27/2024 12:23	1.3	0	0.016	0	1563.213	Mixed
5/27/2024 12:23	1.4	0	0.013	0	1300.854	Mixed
5/27/2024 12:23	1.2	0	0.072	0	7179.783	Mixed
5/27/2024 13:10	0	1.5	0	0	0	Mixed
5/27/2024 13:10	4.3	0	36.187	0.004	3618706.5	WiFi/Phone
5/27/2024 13:10	2.6	0	0.208	0	20825.604	Mixed
5/27/2024 13:10	2.5	0	4.281	0	428090.625	WiFi/Phone
5/27/2024 13:10	3.7	1.1	48.119	0.005	4811894.5	WiFi/Phone
5/27/2024 13:11	4.2	0	108.022	0.011	10802159	WiFi/Phone
5/27/2024 13:11	4.6	0	0.444	0	44383.266	Power Line
5/27/2024 13:11	4	0	1.48	0	148028.297	WiFi/Phone
5/27/2024 13:11	4.2	0	10.151	0.001	1015116.75	WiFi/Phone
5/27/2024 13:11	4.4	0	1.741	0	174133.844	WiFi/Phone
5/27/2024 13:11	4.3	0	1.896	0	189588.984	WiFi/Phone
5/27/2024 13:11	9.5	0	1.896	0	189588.984	WiFi/Phone
5/27/2024 13:11	9.5	0	1.896	0	189588.984	WiFi/Phone
5/27/2024 13:11	9.5	0	1.896	0	189588.984	WiFi/Phone
5/27/2024 13:11	9.5	0	1.896	0	189588.984	WiFi/Phone
5/27/2024 13:11	3.7	0	1.669	0	166883.578	WiFi/Phone
5/27/2024 13:11	3.6	0	0.221	0	22105.422	WiFi/Phone
5/27/2024 13:11	3.4	0	0.363	0	36345.359	WiFi/Phone
5/27/2024 13:11	2.4	0	20.308	0.002	2030764.25	WiFi/Phone
5/27/2024 13:11	2.4	0	2.04	0	204031.75	WiFi/Phone
5/27/2024 13:11	2.7	0	2.056	0	205614.406	WiFi/Phone
5/27/2024 13:11	2.4	0	2.967	0	296731.813	WiFi/Phone
5/27/2024 13:11	2.4	0	2.967	0	296731.813	WiFi/Phone
5/27/2024 13:11	2.4	0	3.23	0	323024.5	WiFi/Phone
5/27/2024 13:11	2.5	0	2.623	0	262252.563	WiFi/Phone
5/27/2024 13:11	2.4	0	2.705	0	270478.313	WiFi/Phone
5/27/2024 13:11	2.3	0	2.145	0	214540.328	WiFi/Phone
5/27/2024 13:11	2.4	0	2.653	0	265307.563	WiFi/Phone
5/27/2024 13:11	2.4	0	2.726	0	272574.625	WiFi/Phone
5/27/2024 13:11	2.3	0	0.033	0	3316.811	WiFi/Phone
5/27/2024 13:11	2.4	0	0.002	0	184.528	WiFi/Phone
5/27/2024 13:11	2.4	0	0.173	0	17331.617	WiFi/Phone
5/27/2024 13:11	2.4	0	0.467	0	46697.324	WiFi/Phone
5/27/2024 13:11	2.4	0	0.432	0	43184.391	WiFi/Phone
5/27/2024 13:11	2.2	0	0.078	0	7777.852	WiFi/Phone
5/27/2024 13:11	2.3	0	0.248	0	24799.494	WiFi/Phone
5/27/2024 13:11	2.3	0	0.267	0	26734.885	WiFi/Phone
5/27/2024 13:11	2.2	0	0.137	0	13667.314	WiFi/Phone
5/27/2024 13:11	2.2	0	0.027	0	2730.218	WiFi/Phone
5/27/2024 13:11	2.1	0	0.09	0	8963.629	WiFi/Phone
5/27/2024 13:11	2.2	0	4.215	0	421537.219	WiFi/Phone
5/27/2024 13:11	2.2	0	4.937	0	493743	WiFi/Phone
5/27/2024 13:11	2.2	0	0.065	0	6484.255	WiFi/Phone
5/27/2024 13:11	2.2	0	4.215	0	421537.219	WiFi/Phone
5/27/2024 13:11	2.1	0	3.627	0	362663.406	WiFi/Phone
5/27/2024 13:11	2.1	0	4.264	0	426442.813	WiFi/Phone
5/27/2024 13:11	2	0	5.053	0.001	505298.625	WiFi/Phone
5/27/2024 13:11	2.2	0	1.293	0	129277.156	WiFi/Phone
5/27/2024 13:11	2.2	0	1.612	0	161175.734	WiFi/Phone
5/27/2024 13:11	2.1	0	3.516	0	351641.406	WiFi/Phone
5/27/2024 13:11	2.1	0	3.516	0	351641.406	WiFi/Phone
5/27/2024 13:11	2.1	0	3.516	0	351641.406	WiFi/Phone
5/27/2024 13:11	2.1	0	3.516	0	351641.406	WiFi/Phone
5/27/2024 13:11	2	0	19.316	0.002	1931583	WiFi/Phone
5/27/2024 13:11	2	0	19.39	0.002	1939037.375	WiFi/Phone
5/27/2024 13:11	2.1	0	11.527	0.001	1152746.75	WiFi/Phone
5/27/2024 13:11	2	0	0.631	0	63054.711	WiFi/Phone
5/27/2024 13:11	2.3	0	0.019	0	1909.735	WiFi/Phone
5/27/2024 13:11	2.4	0	2.064	0	206410.328	WiFi/Phone
5/27/2024 13:11	2.3	0	12.89	0.001	1289006.875	WiFi/Phone
5/27/2024 13:11	2.1	0	1.803	0	180298.594	WiFi/Phone
5/27/2024 13:11	2.2	0	1.234	0	123408.289	WiFi/Phone
5/27/2024 13:11	2.3	0	1.215	0	121511.484	WiFi/Phone
5/27/2024 13:11	2.1	0	1.293	0	129277.156	WiFi/Phone
5/27/2024 13:11	2.1	0	1.099	0	109869.453	WiFi/Phone
5/27/2024 13:12	2.2	0	0.011	0	1145.438	WiFi/Phone
5/27/2024 13:12	2.3	0	0.01	0	989.037	WiFi/Phone
5/27/2024 13:12	2.3	0	0.003	0	304.37	WiFi/Phone
5/27/2024 13:12	2.2	0	0.052	0	5203.372	WiFi/Phone
5/27/2024 13:12	2.2	0	0.012	0	1225.278	WiFi/Phone
5/27/2024 13:12	2.3	0	0.001	0	100.267	WiFi/Phone
5/27/2024 13:12	2.3	0	0.01	0	1008.66	WiFi/Phone
5/27/2024 13:12	2.1	0	0.018	0	1840.335	WiFi/Phone
5/27/2024 13:12	2	0	0.006	0	640.627	WiFi/Phone
5/27/2024 13:12	2.1	0	0.004	0	372.311	WiFi/Phone
5/27/2024 13:12	2	0	0.001	0	64.132	WiFi/Phone
5/27/2024 13:12	2.1	0	0.005	0	522.293	WiFi/Phone
5/27/2024 13:12	2.1	0	0.013	0	1300.854	WiFi/Phone
5/27/2024 13:12	2	0	0.001	0	100.267	WiFi/Phone
5/27/2024 13:12	2	0	0.001	0	100.267	WiFi/Phone
5/27/2024 13:12	2	0	0.001	0	100.267	WiFi/Phone
5/27/2024 13:12	2	0	0.001	0	100.267	WiFi/Phone
5/27/2024 13:12	2	0	0.523	0	52296.508	WiFi/Phone
5/27/2024 13:12	2.3	0	1.298	0	129778.586	WiFi/Phone
5/27/2024 13:12	2.2	0	8.437	0.001	843704.125	WiFi/Phone
5/27/2024 13:12	2.2	0	4.248	0	424801.313	WiFi/Phone
5/27/2024 13:12	2.1	0	9.581	0.001	958110	WiFi/Phone
5/27/2024 13:12	2.1	0	9.581	0.001	958110	WiFi/Phone


I do have 3 more minutes worth of logs just before that if needed.

Is that "0 mixed" line in the log above where i restarted it maybe? does a restart reflect on the log? that would be good to know.

My question is, could this 1000 "XTREME" be a real signal from the powerline, or firmware that needs to be updated, or some other phenomenon? Is the meter working properly or was this a glitch maybe from static? Can you see the extreme signals on the logs above? I'm not sure what to expect from the numbers so I cant tell.
Unfortunately it was when we went to look at the house that we already have an offer on. So now im really curious since this signal well into the range of some of the outbuildings.

Regards



EDIT:
I guess i should add the fact that we heard the sparking that goes on with the transmission lines, after all was said and done, but not necessarily before it spiked. So I wonder if maybe the EMF increases when the lines are under heavy load and it had just went under a heavy load, where the 1.5 was a more-or-less standby current? I was wondering if it was an additional static spike somehow. I was holding the meter in my hand properly by the power portion of the sides and didn't rub it or anything. Anybody care to comment on that if EMF increases when load is higher on transmission lines? FYI, there was 3 lines, the dual t braced wood pole tower wasn't extremely high, there was 3 wires, side-by-side, unstacked, and there was plenty of ceramics, maybe 2.5-3 feet worth, so im pretty sure its a transmission vs. local distribution line. Im my normal testing of other similar looking lines its usually about ~1.5uT under them.
1   L A T E S T    R E P L I E S    (Newest First)
EmfDev Posted - 06/03/2024 : 09:44:14
The RF reading is not high but maybe noise that came from the power lines?