GQ Electronics Technical Support Forum Active Users: / Visits Today:
Highest Active Users:
GQ Electronics Technical Support Forum
Home | Profile | Register | Active Topics | Members | Search | FAQ
Username:
Password:
Save Password
Forgot your Password?

 All Forums
 GQ Electronics Forums
 2.GQ Geiger Muller Counter
 GMC-320 Plus high counting

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
Erfanra Posted - 06/22/2017 : 07:34:25
I have a GMC-320 Plus Geiger counter and I have tested with radioactive sample and it shows about 4000-7000 cpm when it is connected to the charger or pc via usb cable and when i disconnect the cable then it shows 700cpm with same sample, what is the problem?
anybody have this problem?
And is it normal that it takes about 2 minutes until it counts down and reduces 5000CPM to normal background radiations ?

Er
1   L A T E S T    R E P L I E S    (Newest First)
ullix Posted - 06/25/2017 : 00:31:46
I got such significant discrepancies in count rates only with my light-sensitive tube. See the in-depth discussion here: http://www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=4540

Follow that post to verify that your tube is NOT light-sensitive!

CPM=7000 is not lethal but also not negligible ( see calculation here http://www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=4568) . Are you sure that your source could even give off such a high level of radiation? What kind is the source? Is it beta or gamma?

The tube itself adapts to a new radiation level within a millisecond or less, because that is how long a single pulse lasts. But in CPM mode the counter sums up all the events over the past minute, and hence it should adapt to a new rate level within exactly 1 minute. Compare with picture for a change from 550CPM to background. If it takes longer, then something else is going on.