| Reply #1
                 ullix 
                      
                 
                
                Germany 
                1222 Posts  | 
                
                  
                    
                      
                       Posted - 06/25/2017 :  00:31:46
                        
                        
                      
  | 
                     
                    
                       I got such significant discrepancies in count rates only with my light-sensitive tube. See the in-depth discussion here:  http://www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=4540
  Follow that post to verify that your tube is NOT light-sensitive!
  CPM=7000 is not lethal but also not negligible ( see calculation here http://www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=4568) . Are you sure that your source could even give off such a high level of radiation? What kind is the source? Is it beta or gamma?
  The tube itself adapts to a new radiation level within a millisecond or less, because that is how long a single pulse lasts. But in CPM mode the counter sums up all the events over the past minute, and hence it should adapt to a new rate level within exactly 1 minute. Compare with picture for a change from 550CPM to background. If it takes longer, then something else is going on.
  Image Insert:
    29964 bytes | 
                     
                    
                        | 
                     
                   
                 |