GQ Electronics Technical Support Forum Active Users: / Visits Today:
Highest Active Users:
GQ Electronics Technical Support Forum
Home | Profile | Register | Active Topics | Members | Search | FAQ
Username:
Password:
Save Password
Forgot your Password?

 All Forums
 GQ Electronics Forums
 2.GQ Geiger Muller Counter
 M4011 DATASHEET

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
Callisto Posted - 10/25/2012 : 05:29:14
where can I find the datasheet of the M4011 ?
Measurements of the my GMC 300 are not consistent with actual measurements
4   L A T E S T    R E P L I E S    (Newest First)
ZLM Posted - 03/11/2016 : 11:23:07
For the tube noise 0.2cps means average. The true value can be 0 - 24. The calibration process should cover this variation.

Nudnik Posted - 03/11/2016 : 03:55:08
According to government sources we have a background of 77nSv/h ( h**p://sfws.lfrz.at/)
According to the data sheet tube noise is at 0.2cps which is 12cpm.

At the given background radiation of 77nSv/h my GMC320 shows 19cpm.
Now, if I take the tube noise into consideration this would mean 19 - 12 = 7cpm = 77nSv/h

So, either tube noise data are wrong or the tube is crap (at least unusable for background measurement)

Am I missing something?
Craig Posted - 01/21/2015 : 21:36:37
Hi Callisto and Benezi.

Yes: that should be the very first question.

By Googling M4011 + specifications, you end up with incomplete specifications such as at:
> h**p://m.aliexpress.com/item-desc/1790161282.html

--and GQ's own posted specs at:
> h**p://www.ebay.com/itm/New-High-Sensitivity-M4011-Geiger-Counter-Tube-/260835055100?hash=item3cbafb15fc&pt=LH_DefaultDomain_0#ht_928wt_862

* We get a noise figure of "0.2cps" = 12cpm (which no one wants to talk about), but nowhere do we see the stated gamma sensitivity. I see companies which use LND G-M tubes stating sensitivity in a calibrated 1000uR/hr field of cesium-137 radiation --which lobs the count way above any contested levels of internal noise and/or local "background" levels. Those of us doing low level background monitoring then interpret (say) a spec of 3500cpm at 1000uR/hr --to meaningfully indicate 10 uR/hr (0.10 uSv/hr) at 35cpm.

* For what it's worth, using GeigerGraph-4.3 software and the GQ supplied USB adapter cable, I carefully compared the response of a GMC-200 to a Medcom "Inspector" Geiger counter --in a typical (one meter above the ground) outdoor field of background radiation here --and came up with 16cpm at 0.10 uSv/hr --for what I take to be gamma --but who knows what percentage is actually cosmic/muons and internal G-M tube noise? At least that 16cpm should put you on the same page with everyone else --confused as we all might be about what we're actually counting.

* It appears that GQ Electronics software sets an M4011 tubes response to 10 uR/hr at 20cpm. (I don't think the difference is a big deal.)


Craig

Benezi Posted - 11/12/2014 : 14:51:43
A datasheet of the M4011 would help a lot!
Does anyone knows a source?

Thanks a lot,
Benedikt