Fireworx

8 Posts

Posted - 05/26/2018 :  09:16:19  Show Profile  Reply with Quote
Greetings.
I have recently acquired a 500+. What settings constitute having it on it's most sensitive setting?

Thank you.
Reply #1

EmfDev

2250 Posts

Posted - 05/29/2018 :  08:29:48  Show Profile  Reply with Quote
I don't think you can change the sensitivity of the tubes.
Go to Top of Page
Reply #2

Fireworx

8 Posts

Posted - 05/31/2018 :  03:24:48  Show Profile  Reply with Quote
I was read that it has two tubes- one designed for measuring lower levels, and the other for measuring higher levels- like those that would be encountered in an industrial environment. If that is the case, I want put it into the more sensitive mode for measuring lower levels.
Go to Top of Page
Reply #3

EmfDev

2250 Posts

Posted - 05/31/2018 :  09:46:31  Show Profile  Reply with Quote
Those two tubes have different sensitivity but use same high voltage source. Increasing the voltage may increase the sensitivity but will reduce the tube lifespan. There is a potentiometer on the board which can be adjusted to change the voltage of the tubes.
Go to Top of Page
Reply #4

EmfDev

2250 Posts

Posted - 06/11/2018 :  08:32:47  Show Profile  Reply with Quote
@Fireworx. Can you please repost your last response? Thanks.
Go to Top of Page
Reply #5

Fireworx

8 Posts

Posted - 06/15/2018 :  08:15:35  Show Profile  Reply with Quote
Ok. Thanks you EmfDev. I was really just asking if there was a button to press which. My question was based on my assumptions that:

1) the tubes function independently and
2) that the meter may be sold in the default mode of being set to the less sensitive of the tubes.

If your response turns out to be that both the tubes function simultaneously, then that will be the end of my enquiry.
Go to Top of Page
Reply #6

EmfDev

2250 Posts

Posted - 06/15/2018 :  10:27:07  Show Profile  Reply with Quote
The software is not able to do that. The only way is to remove one of the tubes to independently use only one of them. The CPM is sill true from the tube but the usv/h and mr/h need to be calibrated. The low sensitivity tube cannot detect low level radiation and the resolution is going to be bad. Also if you remove the low sensitivity tube, the cpm will be the same as a GMC-500 unit but the usv/h and mr/h need to be calibrated.
Go to Top of Page
Reply #7

nsarmiento

Nicaragua
1 Posts

Posted - 07/02/2019 :  14:29:35  Show Profile  Reply with Quote
I NEED TO CALIBRATE MY 500+
WHICH SOFTWARE DO I USE?

nhvp
Go to Top of Page
Reply #8

EmfDev

2250 Posts

Posted - 07/02/2019 :  15:25:48  Show Profile  Reply with Quote
Hi nsarmiento, the 500+ is already calibrated before we ship it to customers. May I know why would you want to re-calibrate it?

Edited by - EmfDev on 07/02/2019 15:26:09
Go to Top of Page
  Previous Topic Topic Next Topic  
 New Topic  Reply to Topic
 Printer Friendly
Jump To:
GQ Electronics Technical Support Forum © Copyright since 2011 Go To Top Of Page
Generated in 0.09 sec. Snitz's Forums 2000