Benezi

15 Posts

Posted - 12/05/2014 :  12:15:47  Show Profile  Reply with Quote
Hello friends!

I read a few posts here about Maximum Reading of the GMC 280 and 300.
Is it true that the 320+ can display up to 655xx CPM?
But is it possible to have higher measurements by using the logging function?

Because i have one entry in the log file with 2000 CPS. That would be 120000 CPM, much more than the 655xx?

Thanks a lot for your help,
Benedikt
Reply #1

ZLM

1261 Posts

Posted - 12/06/2014 :  20:06:25  Show Profile  Reply with Quote
Yes.

The GMC-320 uses 16 bit counter for both CPM and CPS.

But the display only for CPM, you only can see the 16 bit maximum CPM is 65535.

Since the logging unit is CPS, so the logging has much higher CPM you can get. Theoretically, the logging you can get 65535*60 = 393180 CPM.
Go to Top of Page
Reply #2

Benezi

15 Posts

Posted - 12/07/2014 :  09:22:26  Show Profile  Reply with Quote
Thanks a lot.
Great device!

I'm working on a GMC 320+ with the big SBT10a, which has a conversion factor about 2500. Thats why i asked.

Regards from Germany,
Benedikt
Go to Top of Page