GQ Electronics Technical Support Forum Active Users: / Visits Today:
Highest Active Users:
GQ Electronics Technical Support Forum
Home | Profile | Register | Active Topics | Members | Search | FAQ
Username:
Password:
Save Password
Forgot your Password?

 All Forums
 GQ Electronics Forums
 2.GQ Geiger Muller Counter
 Temperature Sensitivity of Glass Geiger Tubes

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
ullix Posted - 09/22/2022 : 06:01:16
The environment can be so mean to Geiger counters, first it is the light, now it is the temperature.

I have tested my next GeigerLog version using a Raspi as a WiFiServer. The Raspi was serving counts from a GMC-500+, and temperature from an I2C sensor, all via WiFi. Counter and devices were put at a place in my home, where the temperature changes at least 10 °C over the day (near the window), measuring background. This is what happens on the 3 day recording:




Blue is the Geiger counts (left scale), red is the temperature in °C (right scale). The thin double yellow lines mark the Moving Average of the counts.

It is easy to see: the counts are not constant, they move up and down, and the Moving Average is well anti-correlated with the temperature.

GeigerLog offers the "Scat" function, which draws a scatter plot. Here for Temp on the X-axis, and CPM on the Y-axis, with a linear fit as a red line:



There is a very clear correlation. Where does it come from?

Daily flybys of alien spaceships?

As the pressure inside the tube will increase with temperature (as the ideal gas law says), I'd be not surprised if this pressure change results in changing count rates.

But is the tube really to blame, or is it perhaps the electronics of the counter? The counter electronics needs to discriminate count pulses from electric noise pulses, and this may be depending on temperature?