 2.GQ Geiger Muller Counter
 GMC-600 Sensitivity

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
Kiwi-Rob Posted - 11/05/2018 : 00:49:50
Hi guys,

Im pretty new on the scene. I bought a GMC-600 as a gift for my wife as she has always wanted to go to Chernobyl, i am surprising her with a tour there (sssshhh, don't tell her!).
After reading up a little about Geiger Counters i thought that i better test this one before handing it over....
I've used it a couple of days now but no matter where i am the CPM never goes higher then 13. and this is only when i place it behind my running laptop.
I have tried it around my WIFI modem, the microwave, tv, all the things that i figured would give a noticeable reading.
The standard background reading im getting is between 3 and 8 CPM.
Isnt this a bit low?
Im wondering if there is a sensitivity issue?
Is there any way to adjust the sensitivity? How do you guys have your set up?

Im starting to think maybe i should have gotten the 600 plus with the better sensor.

Any help would be greatly appreciated

Thanks!
40   L A T E S T    R E P L I E S    (Newest First)
Searinox Posted - 10/04/2019 : 09:49:25
I got enough check sources including one around 100 uSv/h, unfortunately I have nothing with a known certified output in order to calibrate the device's CPM to uSv/h ratio. Calibration source lab discs will not ship to my country also.
PKIDelirium Posted - 10/03/2019 : 19:11:19
A nice check source is old pre-70's Fiestaware dishes, particularly the Radioactive Red color. You can buy pieces of it on eBay. They got the color by mixing uranium oxide into the glaze.
EmfDev Posted - 10/03/2019 : 09:49:57
This is just an estimate by maybe 2 or 3%.
Searinox Posted - 10/03/2019 : 09:25:11
@EmfDev I should be within reasonable voltage with 20% then as the SBT-11 specs say 400 is recommended operating voltage. Tell me, what is the recommended voltage % increase per 500 CPS for this power supply? I do have a ~30000 cpm source I can test that setting with. Also, getting a voltmeter soon to check actual tube voltage.
EmfDev Posted - 10/03/2019 : 09:13:56
The firmware with 17% default changes by ~20V for every percent and is not the same with the other firmware with 40.00%.
Stargazer 40 Posted - 10/03/2019 : 05:01:59
I have the 600 with an SBT-11A tube. It is recent 6.1 PCB and Auto Voltage Adjust firmware so I see readout as percentage and voltage. The power supply components are the same though and the variability in settings is broad. If GQ had it set to 17% to get proper voltage, then a move to 19% should only raise the voltage by about 13V. I don't think that should even get close to taking it out of GM effect much less out of the plateau region of the tube. You probably should email support and see if they see a problem with power supply since it happens with two different tubes.
Searinox Posted - 10/03/2019 : 00:57:43
@Stargazer_40 What device model did you test that on? My GMC-600 came with 17% default voltage and upping its stock tube beyond 19% starts sending it in avalanche and recording much higher CPM in background relative to actual samples. The same happens at around 28% with the other tube I am testing and that's way below 40%, which I'm not testing out of fear of damaging the tubes as I'm pretty sure by that point it's all just one continuous spark.
Sonicmixmaster Posted - 10/02/2019 : 17:56:58
I don't know if it was mentioned in the previous posts but I have found old ceramic tiles in my friends bathroom to be highly radioactive that you can test. Check this post http://www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=5395
You can also try some ceramic pots or even old ceramic dishes.
Stargazer 40 Posted - 10/02/2019 : 14:17:04
Averages about 6.5v per percentage point. For my 600 at 40% I have 405v. At 42% I have 418v.
Searinox Posted - 10/02/2019 : 13:32:38
Does anyone know how many volts per percentage the firmware settings mean? As I go beyond 20% I start having issues with the tube. Touching the metallic protective grid gives a continuous speaker beep and registers ridiculous values during that. It's "moody" and only happens at times but it worsens with voltage increase until by ~27% it will almost always do this when the grid is touched.

I opened the device and checked, the grid isn't touching any wiring, I even placed insulating tape around the tube's window to make sure it isn't touching the tube's metal shell anywhere. It also only ever seems to do it when I'm touching this with my finger and not anything else so my guess is it's some weird induction or electrostatic event and as soon as I can find a rigid enough replacement mesh that is not metal, I'll swap it out but for now I just wanna know voltage to % conversion to see if I am indeed going overboard.
Stargazer 40 Posted - 09/29/2019 : 02:10:26
@Searinox - I agree with you on mixed bag regarding the SBT-11/A. A thinner mica window may not make a difference. I need to check the specs on the current draw. I have described the design as inefficient in the Auto Voltage Adjust thread, and I think we have those numbers on the spec sheet. GQ does not really differentiate between the two. I will say though that without an alpha source there is no real way to assess the window thickness variation's effect. I have both, but haven't taken the time to swap them in the 600. Another experiment for another day.
Searinox Posted - 09/28/2019 : 14:13:46
I can confirm the sensitivity difference between batches overriding 11 vs 11A sensitivity in some cases. My device came with a 11A tube and I have a source which I could reliably test with that whenever the device sat on top of it, it would register 20 uSv/h. Tested by switching it with a regular 11 tube later and it was picking up 23 uSv/h with the same source and setup(same CPM ratio, same tube voltage%). The source was beta and gamma only so I cannot vouch for how it affected alpha but it's reasonable to imagine a tube with better internal sensitivity and a thicker window sometimes producing better results than a thinner window but more insensitive tube.
PKIDelirium Posted - 09/14/2019 : 07:07:53
quote:
Originally posted by Stargazer 40

@ullix, the 600 is a mica windowed alpha tube like the 7317. There is a grated cutout on the bottom of the case for both the 7317 and the SNT-11/11A. Judicious use of a different GQ label to cover the opening not needed leaves the other open for tube to receive alphas. I still like the idea of a clip on steel beta shield that covers the bottom of the meter and both opening positions.



Like the back “door” on the RadiaScan? 8:33 in the video: https://youtu.be/T1HZ8NAXu64
Damien68 Posted - 09/12/2019 : 23:58:29
@Scutti,
After you have to take a photo and post it here :)
Scutti Posted - 02/15/2019 : 04:23:50
Testing the device: go visit a flea market and look out for old clocks from the 50s-60s (radium painted dials) and yellow-greenish vases (uranium glass). Just turn off the clicks to remain incognito :)
rolex452 Posted - 01/05/2019 : 17:00:54
Re: the SBT11/SBT11A controversy: here is an ad excerpt from an internet seller of computer kits (located in Lithuania)...I hope this helps:
A little hint regarding difference between SBT-11 without A, and SBT-11A: Both tubes are absolutely same alpha capable tubes with the same sensitivity (in fact my tests show that SBT11 might even be a little bit more sensitive than SBT-11A depending on batch and condition). The difference is that at 55C (131F) and higher temperatures the inherent background of the SBT-11 2.5 pulses/s while for SBT-11A it is 0.67 pulses/s. Inherent background means how many false pulses (which are not the result of the particle detection) the tube generates - yes Geiger-Muller tubes most evidently pancake ones have such a little "sin" - but who hasn't. At room temperatures 25C (77F) the inherent background of tubes is the same (0.25 pulses/second).
EmfDev Posted - 12/03/2018 : 14:31:49
@Kiwi-Rob,
According to support, they are still following up on the return.

@ullix, @Stargazer_40
I think this is installation issue, maybe the +/- terminals of the tube are reversed.
Stargazer 40 Posted - 12/02/2018 : 04:01:30
@ullix, the 600 is a mica windowed alpha tube like the 7317. There is a grated cutout on the bottom of the case for both the 7317 and the SBT-11/11A. Judicious use of a different GQ label to cover the opening not needed leaves the other open for tube to receive alphas. I still like the idea of a clip on steel beta shield that covers the bottom of the meter and both opening positions.

Image Insert:

77113 bytes
Stargazer 40 Posted - 12/02/2018 : 04:00:50
@ullix, the 600 is a mica windowed alpha tube like the 7317. There is a grated cutout on the bottom of the case for both the 7317 and the SNT-11/11A. Judicious use of a different GQ label to cover the opening not needed leaves the other open for tube to receive alphas. I still like the idea of a clip on steel beta shield that covers the bottom of the meter and both opening positions.
ullix Posted - 12/02/2018 : 00:26:14
Oh, I missed the fact that the 600 has some sort of a window built-in; thus removing the backplate is not as relevant as in the M4011 based counters.

A higher count rate in this new device compared to your previous is a miracle. Unless the voltage setting was way too low?
Stargazer 40 Posted - 12/01/2018 : 07:18:12
Nice to hear that this is now meeting your expectations. I hope you post some screen shots from your visit to Chernobyl.
Kiwi-Rob Posted - 12/01/2018 : 04:24:15
@ullix:
Thanks for your feedback! After talking with GQ Support (John) We came to the conclusion that i should send the unit back for a refund and order a new one. The new unit arrived and it works brilliantly! A lot more sensitive and can easily get a reading of 150CPM from my bag of KCI. You are correct in thinking that i didnt remove the cover form the first unit, i didnt want to void any guarantee on it. The new unit easily gets readings without removing any covers or "wrapping" the unit in KCI.
Now i am just waiting for my refund for the first unit.... Im hoping GQ are as helpful with their refunds as they are with tech support!
ullix Posted - 11/29/2018 : 04:09:49
@Kiwi-Rob: from what you say I can only conclude that your counter is working as it is supposed to. I don't see an indication for a defect!

As your description does not say that you had removed the back-plate of the counter, I assume you have not.

In that case you are blocking most of the betas, and a count of 48 is even more than my result of CPM=37.16 with a similar setting. Though given this is a Poisson distribution with a StdDev = Squareroot(N), the two are not necessarily different.

Read the Experimental chapters in the Potty Training (https://sourceforge.net/projects/geigerlog/files/GeigerLog-Potty%20Training%20for%20Your%20Geiger%20Counter-v1.0.pdf/download ) again, and repeat the measurements with the backplate removed!

For best performance: Divide your KCl into two plastic bags with at least a pound in each, position the counter on top of the 1st bag, with the tube in close contact to the bag. Put the second bag on top of it; wrap it around the edges a little bit.

The goal is to have a similar amount of KCl all around the counter, e.g. with a thickness of ~2cm.

Then you should get near 150CPM.



EmfDev Posted - 11/12/2018 : 10:19:25
Support will get back to you soon. I think that your unit needs to get fixed and will have latest firmware when you get it back.
Kiwi-Rob Posted - 11/12/2018 : 06:10:29
Yes, I left the meter on the salt for about 30 minutes. I emailed support about the firmware update but was told that there is no update from 2.0 yet. It is still being tested and i will get a copy when it is released.
I have now emailed support again to request repair or replacement of this unit.
Stargazer 40 Posted - 11/12/2018 : 03:33:38
quote:
Originally posted by Kiwi-Rob

So, Thanks for all of the input guys.
My KCL arrived. The maximum reading i am getting is 48 CPM. That is with the unit directly on the bag!




Take a look at my 600 review here, Reply #4

www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=5519

GQ got 110 CPM from potassium salt (from their site), but they were using a substitute salt shaker and not a bag I believe (larger amount spread across the tube window). My 600 11A got 206CPM from the 1 lb. bag. So your GM tube is about 1/4 of what I think it should measure, as whether 11 or 11A, response to the potassium salts should be very similar. Certainly variability in these tubes, but that seems like a lot. I presume you left the meter on the salt for a period of a few minutes to stabilize. Did you upgrade your firmware to 2.1?
Kiwi-Rob Posted - 11/12/2018 : 01:35:18
So, Thanks for all of the input guys.
My KCL arrived. The maximum reading i am getting is 48 CPM. That is with the unit directly on the bag!
Stargazer 40 Posted - 11/08/2018 : 05:40:42
@ullix, I responded with correct translations of a number of sections in #7 above. Your 2nd to last paragraph is tube generated noise pulses in cps, and not mica window thickness.
ullix Posted - 11/08/2018 : 01:58:11
As apparently none of the forum readers is speaking Russian, I fumbled my way through some of the SBT11(A) specs using this side for conversion help: https://en.bab.la/dictionary/russian-english/%D0%BD%D0%BE%D1%80%D0%BC%D0%B0 and annotated the spec-sheet with my discoveries:

Image Insert:

362809 bytes

As is to be expected for spec sheets, the narrow columns contain the lower and upper limits for the properties described in the left column.

Whenever you see "SBT11, SBT11A" separated by a comma and on the same line, the specs given in the two standards column are the lower and upper limits applicable to both tubes. You see that the two tubes are identical in all aspects but one.

This one difference is in the topic literally translated as "Properties Foil". I'd say this is the thickness of the window material (perhaps "Mica" window) given as areal density mg/cm². For other tubes this is typically around 2 mg/cm², so a number of 2.5 for a SBT11 makes sense. The number of 0.67mg/cm² for the SBT11A tube means it is VERY thin! For some perspective: when tubes are additionally offered as "military-grade" their thickness is 3...5mg/cm² instead of 2!

So, when you got an SBT11 (non-A) tube, consider yourself lucky, as you didn't get such a fragile tube as the SBT11A!

The GQ decision to put SBT11 tubes, and not the A-version, into their consumer products is a very reasonable one! I question whether there is even a need to offer a windowed tube. It does allow for marketing claims (perhaps a good enough reason for a company trying to sell products) but I still have not seen anyone demonstrating a measurement of an alpha signal, or a very-low-energy beta signal. Anyone able to show it?



EmfDev Posted - 11/07/2018 : 16:39:18
I hope you get something from the KCL. Let us know about the results thanks!
Kiwi-Rob Posted - 11/07/2018 : 11:22:44
Thank you everyone for sharing your knowledge and helping this newbie out!
It turns out that smoke detectors with Americium here (in Austria) are banned which explains why i wasn't getting anything from the little black box that lives inside my smoke detector (feeling a little like a idiot!).
Im looking forward to seeing what happens with the KCL and will contact gq support if the unit keeps going to 0
Thanks again guys!
EmfDev Posted - 11/07/2018 : 09:55:46
@Kiwi-Rob, the counter shouldn't have 0 CPM. Does that mean the CPM display is 0? I'll have to check with support if you need to return it for replacement/repair.
I think we use the SBT-11 and not the A. Also, WiFi/microwaves/tv shouldn't affect the reading. They emit a different type of radiation (EMF).

Edit. If the display CPM often goes 0, you can contact/email support@gqelectronicsllc.com for repair/replacement. Sorry for the inconvenience.
Stargazer 40 Posted - 11/07/2018 : 05:43:15
quote:
Originally posted by Kiwi-Rob

Hi EmfDev,

Yes, my count often goes to 0.
I tested the unit on a broken smoke detector (covers of smoke detector removed) and got no response. I have ordered some KCL and will run a few tests when it arrives.




Alpha is particle emission and doesn't bounce that I have found. The source in the smoke detector is in a metal cup and it has a metal dome over it that is (normally) plastic clipped and two tabs of the dome soldered to the PCB. When you remove that metal cap by unclipping and desoldering you should be able to see the small yellowish metal chunk of Americium kind of pressed into a tiny cup in the larger metal cup. Try to get to that point and then put the meter over the cup and see if it doesn't respond. That's normally shown to be less than .8-9uCi and quite active for alpha, but you're not going to see it unless the dome is removed.

As I understand the process for manufacture, they measure thickness of natural mica sheets and at some threshold it's designated A and thicker than that it's not. So even in A's there will be variation.
Stargazer 40 Posted - 11/07/2018 : 05:19:43
@ullix, The second to last paragraph talks about random pulses from device itself. Russian says 'Own source pulse' The sensitivity to Cs-137 source is in para above that. It is in CPS and shows 440CPS for 11 and 490CPS for 11A. Above that it specifically talks about sensitivity to gamma radiation and is in CPM, with 11 at 44CPM and 11A at 49CPM. Relative sensitivity to either total field of some amount of CS-137 or for just gamma radiation is that the 11A is 11% more sensitive (or ~10% as EmfDef notes above).
ullix Posted - 11/07/2018 : 02:05:04
Don't worry too much about the differences SBT11 to SBT11A. Here is a datasheet:

Image Insert:

233863 bytes

The only difference is in the 2nd before last paragraph, and perhaps someone can decipher it?

Even if the 2.50 vs 0.67 (that is what I read) were the thickness of the Mica window, and hence would suggest a different sensitivity to alphas and low energy betas, it has almost no relevance for the things you might do with a Geiger counter for the following reasons:

There is no pure alpha emitter. All of them represent one isotope within a decay chain, and this decay chain then has beta and gamma emitters. You will always be able to detect such a substance based on latter emitters even with M4011 tubes. Discriminating between them is an not-so-easy job. You need to make sufficiently long term successive measurements with blocking for alpha and perhaps even beta, without touching the geometry in the slightest. I believe only GeigerLog could handle such experiments; testing for Poisson will be an important part of it!

Should you ever succeed in one, please present it!

P.S. By the way, you see in the diagram that the SBT11(A) consists of 4 tubes connected in parallel. This brings back the idea that a Geiger counter could be constructed to cover a range of sensitivity by using only 1, 2, 3, 4 or even more tubes. This can already be done with with any SBT11(A) or with a set of SBT10 or SBT21 tubes. Any of the many high voltage NPN or MOSFET transistors could act as a switch within the HV lines. And all could work from a single HV source.

Kiwi-Rob Posted - 11/06/2018 : 23:02:56
Hi EmfDev,

Yes, my count often goes to 0.
I tested the unit on a broken smoke detector (covers of smoke detector removed) and got no response. I have ordered some KCL and will run a few tests when it arrives.
I find it odd and a little disapointing that it is just the "luck of the draw" as to whether you get a 11 or 11A given that you pay the same price.
I havent checked to see what i have yet because i dont want to void any warranty if my unit turns out to have a problem.
So, if i dont get any response from the KCL test then what would the next step be...
Thank you for your help.
EmfDev Posted - 11/05/2018 : 17:10:42
Hi Kiwi-Rob, have your count ever gone down to 0 sometimes? Those readings might look normal but if you didn't get good results
with the KCL or smoke detector (still gets 3-8 CPM) then there might be a problem with your device. Maybe tube installation issue.
The SBT-11A is a little bit more sensitive, maybe 10%, but I'm not that sure about how sensitive it is. Yes it is able to detect alpha, maybe if you have a broken smoke detector, you can test it with that one also.
Kiwi-Rob Posted - 11/05/2018 : 13:17:56
Thanks Guys! I will get some KCI and run some tests.
I have just been reading about the difference between SBT-11 and SBT-11A and the general consensus is that the SBT-11 isnt made for picking up Alpha at all! but GQ are selling these units as being able to pick up Alpha? If the difference between the sensor models is true then why dont GQ care which one they use??
Stargazer 40 Posted - 11/05/2018 : 06:13:53
GQ doesn't really differentiate between SBT-11 and SBT-11A. The SBT-11A has a thinner mica window and will be the more sensitive tube. If your 600 is an 11 then it will be a little less sensitive, but that is really only for alphas and maybe a bit for beta. I'm in a fluvial valley in front of the Pre-Cambrian Rocky Mountains (ie, lots of granite and metamorphic deposition) I see about 25-30CPM background. I have an 11A in my 600. While the 7317 is more sensitive, I can assure you that the 600 will respond more than adequately when you get into a place where counts can go very high. See my 600 review for some of the counts I've been able to achieve with some handy sources.

h**p://www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=5519 (replace the asterics with 't's)

There is no way to adjust sensitivity. Counts is counts, and in a previous thread GQ has said that it takes a bit for the SBT-11 to settle down. Yours appears to be working. You just need (as ullix has suggested) to get a test source with more activity. The meter does respond well to Potassium Salt.


ullix Posted - 11/05/2018 : 02:07:26
Indeed, 3 to 8 CPM does seem to be low. But an SBT-11 tube should be plenty good enough to give you a decent readout.

I had tested the SBT11A tube and found it to be similar or more sensitive to Background, KCl and Gas mantles than the M4011 tube (used in most other GQ counters) and the SBM20 tube.

For easy testing I suggest to use a Kalium (Potassium) salt, like KCl. Available if not in hardware stores then on Amazon, or even in health food stores (as an alternative salt to NaCl). You need a pound at least, better 2 pounds, Should be no more than ~$10. Something like this: https://www.amazon.de/Kaliumchlorid-Di%C3%A4tsalz-1kg-KCl-chem/dp/B00IAJEYBY/
or this health food item https://www.amazon.de/Kaliumchlorid-Kaliumsalz-Lebensmittelqualit%C3%A4t-Di%C3%A4t-Salzersatz/dp/B01N4LKMTD


Read more details and howto-measure in my Potty-Training article, https://sourceforge.net/projects/geigerlog/files/GeigerLog-Potty%20Training%20for%20Your%20Geiger%20Counter-v1.0.pdf/download

With KCl and the procedure described in the article you should get something near 150CPM. If that is not reached, then perhaps the voltage setting is wrong. But don't touch this for now!

If that also is not helping, then there may be a defect of some kind, but try the KCL test first.