Posted - 11/09/2012 : 08:36:38 Routine monitoring using the GMC-300 over two weeks in the Cleveland Ohio area shows 0 CPM occasionally (once every 4-5 days). Is this just a statistical fluke or is there some other reason?
A curvefit of the data is a flat line of 20.8 CPM with a peak of 38 CPM. So I would expect 2 CPM with a similar frequency as the 38 CPM, but since I never get 40 CPM at my site, I expect no 0 CPM. |
3 L A T E S T R E P L I E S (Newest First) |
ZLM |
Posted - 11/15/2012 : 10:43:10 There are some problems on voltage I believe.
You can contact GQ ask for a checking. You may need to return the unit to GQ.
OR If you know the electronics, then try to use a 1000V small (~0.1uf) capacitor, parallel to the tube two ends and see if it changes anything. |
jfin4 |
Posted - 11/09/2012 : 13:32:45 quote: Originally posted by OnTourWithGerrit
Well, have a look inside the Counter and check the electrical connection from the tube to the onboard connections. If this is broken, the Counter is unable to count. I have had the same problem with a another counter.
No, it works fine in that it generates CPM readings with a non-zero value most of the time. But some infrequent 1 minute periods are completely zero. Statistical fluke? |
OnTourWithGerrit |
Posted - 11/09/2012 : 11:41:54 Well, have a look inside the Counter and check the electrical connection from the tube to the onboard connections. If this is broken, the Counter is unable to count. I have had the same problem with a another counter. |