T O P I C R E V I E W |
ska |
Posted - 11/03/2012 : 07:20:39 Hi, I've a problem with my GMC-300 there is a singolarity in uSV/h misure at about 300CPM switch from 0.7uSv/h to 2.00 uSV/h, I reset to factory setting but the problem is still there. Could I know how I can solve the situation?
Regards |
3 L A T E S T R E P L I E S (Newest First) |
ZLM |
Posted - 11/07/2012 : 01:18:53 The problem is the calibration data. In calibration menu, change your last uSv data to 3.75 or close value. |
ska |
Posted - 11/05/2012 : 05:41:09 The problem is that at 240 CPM I have a 0.7 uSv/h as dose rate, at 243 I have 1.91 uSv/h!
My SW version is 2.21
|
ZLM |
Posted - 11/04/2012 : 19:42:51 What is you exact problem? can you upload a picture?
What is your firmware version? |