GQ Electronics Technical Support Forum Active Users: / Visits Today:
Highest Active Users:
GQ Electronics Technical Support Forum
Home | Profile | Register | Active Topics | Members | Search | FAQ
Username:
Password:
Save Password
Forgot your Password?

 All Forums
 GQ Electronics Forums
 2.GQ Geiger Muller Counter
 GMC-500+ Readings Too High Compared to Other Count

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
skylerclooney Posted - 09/05/2025 : 19:43:48
Hi everyone,
I’m Skyler and new here on the GQ forum. I’ve been using a GQ GMC-500+ Geiger Muller Counter for a few weeks, and I’m running into a problem I can’t figure out.
When I compare the readings from my unit with another counter, mine consistently shows values that are about 20–25% higher. I’ve tried resetting the device, updating the firmware, and even checking in different environments, but the discrepancy remains. I also noticed that sometimes the CPM spikes briefly without any apparent radiation source nearby.
Could this be a calibration issue, or something else I might be missing? Do I need to adjust settings manually, or is there a recommended procedure to check accuracy?
Any advice would be greatly appreciated.
1   L A T E S T    R E P L I E S    (Newest First)
ullix Posted - 09/06/2025 : 00:19:09
The "spikes" are probably the result of an inappropriate setting in the counter's firmware. See in particular my Reply #7 in https://www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=10932

Among possible settings, the default value of FET=3 is the worst setting! Switch FET off by setting it to FET=60!

If FET is off, and you still get spikes, they might be real - a consequences of the random nature of radioactivity. You can verify ok or not by running a Poisson check in GeigerLog https://sourceforge.net/projects/geigerlog/

Comparing different counter readings can be tricky. To which other counter(s) are you comparing your GMC-500 readings? Which Geiger tubes are those using?

This same question had been asked very recently, answer in Reply#1 in https://www.gqelectronicsllc.com/forum/topic.asp?TOPIC_ID=10933 .

You did not mention your radio source. Chances are it has significant beta radiation?