GQ Electronics Technical Support Forum Active Users: / Visits Today:
Highest Active Users:
GQ Electronics Technical Support Forum
Home | Profile | Register | Active Topics | Members | Search | FAQ
Username:
Password:
Save Password
Forgot your Password?

 All Forums
 GQ Electronics Forums
 2.GQ Geiger Muller Counter
 Calibration intervals

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
Bronzetower Posted - 09/05/2023 : 00:14:02
I’ve been running the GCM-500+ continuously 24/7 for about 10 months now. It was set up to measure ambient radiation and is logging datapoints every 2 minutes. I’ve noticed that for the first 7 months or so, I never saw any reading above 50 CPM. Now after 8 months I’m seeing 50CPM+ readings two or three times a day.

I’m wondering how frequently these things needs to be recalibrated?
2   L A T E S T    R E P L I E S    (Newest First)
EmfDev Posted - 09/06/2023 : 10:48:06
Try to do a factory reset. Check if the 50+ CPM is consistent even in different areas.
ullix Posted - 09/05/2023 : 03:03:23
"Calibration" is of no help; it is only the introduction of a factor to convert from CPM to microSievert/h.

It looks like hardware defect. Try different USB-cable, check battery, check anode voltage.