Hello,
Just received the GMC-300 as a Christmas present. - Why, I currently live in Japan. My mom has been concerned about me being here since the earthquake and all. Anyway, cool toy anyway I think.
I'm an engineer so pretty up to date with things, but I do have some questions. This is the first day I've had this, and I'm actually seeing some spikes around 200-300CPM at times and it will go on that way for dunno - an hour maybe. Is this normal radiation to be expected anywhere? Really, reeeaally curious. Want to know if there is anything I need to do do calibrate it. It is currently 2.3 firmware.
Would really appreciate any feedback. |
|
| Reply #1
ZLM
    
1271 Posts |
Posted - 12/30/2012 : 11:07:29
|
Different places have different radiation background level. But, I think the most of city background radiation level should not exceed the 100CPM with GMC-300.
Did you see spikes on the unit LCD disply OR you see the data log from computer?
|
 |
|
| Reply #2
fictitious1

4 Posts |
Posted - 12/30/2012 : 16:40:06
|
Here are the numbers I recorded last night using heartbeat mode. I did not include the seconds column because of space. I live in Takasaki, Japan.
Also, how long will it record data to the device in minute mode? Day, week? I would like to switch to second mode but want to know how long I have before I need to download it to the computer to keep a running record.
Thanks
Date Time CPM 12/30/2012 22:16 38 12/30/2012 22:17 39 12/30/2012 22:18 123 12/30/2012 22:19 189 12/30/2012 22:20 213 12/30/2012 22:21 185 12/30/2012 22:22 85 12/30/2012 22:23 127 12/30/2012 22:24 218 12/30/2012 22:25 177 12/30/2012 22:26 201 12/30/2012 22:27 171 12/30/2012 22:28 104 12/30/2012 22:29 226 12/30/2012 22:30 71 12/30/2012 22:31 29 12/30/2012 22:32 43 12/30/2012 22:33 29 12/30/2012 22:34 27 12/30/2012 22:35 39 12/30/2012 22:36 30 12/30/2012 22:37 35 12/30/2012 22:38 27 12/30/2012 22:39 32 12/30/2012 22:40 43 12/30/2012 22:41 41 12/30/2012 22:42 38 12/30/2012 22:43 36 12/30/2012 22:44 31 12/30/2012 22:45 43 12/30/2012 22:46 73 12/30/2012 22:47 150 12/30/2012 22:48 135 12/30/2012 22:49 255 12/30/2012 22:50 217 12/30/2012 22:51 223 12/30/2012 22:52 207 12/30/2012 22:53 186 12/30/2012 22:54 168 12/30/2012 22:55 243 12/30/2012 22:56 239 12/30/2012 22:57 201 12/30/2012 22:58 195 12/30/2012 22:59 225 12/30/2012 23:00 227 12/30/2012 23:01 235 12/30/2012 23:02 135 12/30/2012 23:03 179 12/30/2012 23:04 200 12/30/2012 23:05 146 12/30/2012 23:06 104 12/30/2012 23:07 190 12/30/2012 23:08 154 12/30/2012 23:09 69 12/30/2012 23:10 144 12/30/2012 23:11 182 12/30/2012 23:12 201 12/30/2012 23:13 127 12/30/2012 23:14 191 12/30/2012 23:15 241 12/30/2012 23:16 140 12/30/2012 23:17 131 12/30/2012 23:18 126 12/30/2012 23:19 171 12/30/2012 23:20 250 12/30/2012 23:21 191 12/30/2012 23:22 192 12/30/2012 23:23 327 12/30/2012 23:24 136 12/30/2012 23:25 216 12/30/2012 23:26 150 12/30/2012 23:27 74 12/30/2012 23:28 69 12/30/2012 23:29 40 12/30/2012 23:30 28 12/30/2012 23:31 30 12/30/2012 23:32 37 12/30/2012 23:33 29 12/30/2012 23:34 40 12/30/2012 23:35 28 12/30/2012 23:36 30 12/30/2012 23:37 37 12/30/2012 23:38 32 12/30/2012 23:39 33 12/30/2012 23:40 21 12/30/2012 23:41 42 12/30/2012 23:42 30 12/30/2012 23:43 34 12/30/2012 23:44 51 12/30/2012 23:45 41 12/30/2012 23:46 37 12/30/2012 23:47 22 12/30/2012 23:48 26 12/30/2012 23:49 49 12/30/2012 23:50 48 12/30/2012 23:51 39 12/30/2012 23:52 28 12/30/2012 23:53 62 12/30/2012 23:54 122 12/30/2012 23:55 244 12/30/2012 23:56 135 12/30/2012 23:57 160 12/30/2012 23:58 134 12/30/2012 23:59 154 12/31/2012 0:00 85 12/31/2012 0:01 154 12/31/2012 0:02 256 12/31/2012 0:03 262 12/31/2012 0:04 160 12/31/2012 0:05 146 12/31/2012 0:06 145 12/31/2012 0:07 158 12/31/2012 0:08 167 12/31/2012 0:09 240 12/31/2012 0:10 256 12/31/2012 0:11 243 12/31/2012 0:12 176 12/31/2012 0:13 182 12/31/2012 0:14 181 12/31/2012 0:15 165 12/31/2012 0:16 229 12/31/2012 0:17 108 12/31/2012 0:18 177 12/31/2012 0:19 122 12/31/2012 0:20 120 12/31/2012 0:21 174 12/31/2012 0:22 168 12/31/2012 0:23 100 12/31/2012 0:24 137 12/31/2012 0:25 219 12/31/2012 0:26 126 12/31/2012 0:27 155 12/31/2012 0:28 133 12/31/2012 0:29 201 12/31/2012 0:30 178 12/31/2012 0:31 165 12/31/2012 0:32 230 12/31/2012 0:33 285 12/31/2012 0:34 226 12/31/2012 0:35 170 12/31/2012 0:36 97 12/31/2012 0:37 222 12/31/2012 0:38 154 12/31/2012 0:39 181 12/31/2012 0:40 146 12/31/2012 0:41 105 12/31/2012 0:42 93 12/31/2012 0:43 245 12/31/2012 0:44 198 12/31/2012 0:45 202 12/31/2012 0:46 256 12/31/2012 0:47 206 12/31/2012 0:48 143 12/31/2012 0:49 166 12/31/2012 0:50 187 12/31/2012 0:51 98 12/31/2012 0:52 138 12/31/2012 0:53 161 12/31/2012 0:54 196 12/31/2012 0:55 153 12/31/2012 0:56 171 12/31/2012 0:57 213 12/31/2012 0:58 140 12/31/2012 0:59 35 12/31/2012 1:00 39 12/31/2012 1:01 48 12/31/2012 1:02 154 12/31/2012 1:03 180 12/31/2012 1:04 153 12/31/2012 1:05 296 12/31/2012 1:06 142 12/31/2012 1:07 200 12/31/2012 1:08 97 12/31/2012 1:09 160 12/31/2012 1:10 217 12/31/2012 1:11 169 12/31/2012 1:12 178 12/31/2012 1:13 152 12/31/2012 1:14 101 12/31/2012 1:15 193 12/31/2012 1:16 170 12/31/2012 1:17 258 12/31/2012 1:18 178 12/31/2012 1:19 156 12/31/2012 1:20 193 12/31/2012 1:21 138 12/31/2012 1:22 231 12/31/2012 1:23 137 12/31/2012 1:24 125 12/31/2012 1:25 219 12/31/2012 1:26 213 12/31/2012 1:27 291 12/31/2012 1:28 179 12/31/2012 1:29 168 12/31/2012 1:30 175 12/31/2012 1:31 133 12/31/2012 1:32 171 12/31/2012 1:33 124 12/31/2012 1:34 211 12/31/2012 1:35 127 12/31/2012 1:36 249 12/31/2012 1:37 164 12/31/2012 1:38 216 12/31/2012 1:39 161 12/31/2012 1:40 278 12/31/2012 1:41 181 12/31/2012 1:42 184 12/31/2012 1:43 168 12/31/2012 1:44 227 12/31/2012 1:45 238 12/31/2012 1:46 148 12/31/2012 1:47 35 12/31/2012 1:48 49 12/31/2012 1:49 39 12/31/2012 1:50 45 12/31/2012 1:51 26 12/31/2012 1:52 30 12/31/2012 1:53 24 12/31/2012 1:54 46 12/31/2012 1:55 35 12/31/2012 1:56 115 12/31/2012 1:57 123 12/31/2012 1:58 114 12/31/2012 1:59 148 12/31/2012 2:00 193 12/31/2012 2:01 235 12/31/2012 2:02 237 12/31/2012 2:03 192 12/31/2012 2:04 229 12/31/2012 2:05 198 12/31/2012 2:06 116 12/31/2012 2:07 182 12/31/2012 2:08 157 12/31/2012 2:09 262 12/31/2012 2:10 99 12/31/2012 2:11 214 12/31/2012 2:12 177 12/31/2012 2:13 133 12/31/2012 2:14 192 12/31/2012 2:15 199 12/31/2012 2:16 190 12/31/2012 2:17 195 12/31/2012 2:18 251 12/31/2012 2:19 183 12/31/2012 2:20 154 12/31/2012 2:21 178 12/31/2012 2:22 255 12/31/2012 2:23 165 12/31/2012 2:24 309 12/31/2012 2:25 266 12/31/2012 2:26 187 12/31/2012 2:27 251 12/31/2012 2:28 191 12/31/2012 2:29 178 12/31/2012 2:30 105 12/31/2012 2:31 273 12/31/2012 2:32 140 12/31/2012 2:33 178 12/31/2012 2:34 230 12/31/2012 2:35 205 12/31/2012 2:36 186 12/31/2012 2:37 191 12/31/2012 2:38 187 12/31/2012 2:39 210 12/31/2012 2:40 210 12/31/2012 2:41 204 12/31/2012 2:42 222 12/31/2012 2:43 210 12/31/2012 2:44 180 12/31/2012 2:45 239 12/31/2012 2:46 162 12/31/2012 2:47 286 12/31/2012 2:48 230 12/31/2012 2:49 212 12/31/2012 2:50 198 12/31/2012 2:51 148 12/31/2012 2:52 186 12/31/2012 2:53 182 12/31/2012 2:54 225 12/31/2012 2:55 184 12/31/2012 2:56 209 12/31/2012 2:57 183 12/31/2012 2:58 227 12/31/2012 2:59 352 12/31/2012 3:00 190 12/31/2012 3:01 215 12/31/2012 3:02 228 12/31/2012 3:03 229 12/31/2012 3:04 161 12/31/2012 3:05 212 12/31/2012 3:06 223 12/31/2012 3:07 281 12/31/2012 3:08 103 12/31/2012 3:09 306 12/31/2012 3:10 303 12/31/2012 3:11 151 12/31/2012 3:12 209 12/31/2012 3:13 267 12/31/2012 3:14 204 12/31/2012 3:15 263 12/31/2012 3:16 174 12/31/2012 3:17 203 12/31/2012 3:18 218 12/31/2012 3:19 200 12/31/2012 3:20 210 12/31/2012 3:21 156 12/31/2012 3:22 115 12/31/2012 3:23 256 12/31/2012 3:24 205 12/31/2012 3:25 239 12/31/2012 3:26 192 12/31/2012 3:27 256 12/31/2012 3:28 246 12/31/2012 3:29 137 12/31/2012 3:30 209 12/31/2012 3:31 145 12/31/2012 3:32 204 12/31/2012 3:33 149 12/31/2012 3:34 213 12/31/2012 3:35 230 12/31/2012 3:36 147 12/31/2012 3:37 272 12/31/2012 3:38 187 12/31/2012 3:39 198 12/31/2012 3:40 129 12/31/2012 3:41 213 12/31/2012 3:42 198 12/31/2012 3:43 291 12/31/2012 3:44 275 12/31/2012 3:45 345 12/31/2012 3:46 266 12/31/2012 3:47 143 12/31/2012 3:48 142 12/31/2012 3:49 224 12/31/2012 3:50 246 12/31/2012 3:51 200 12/31/2012 3:52 170 12/31/2012 3:53 276 12/31/2012 3:54 212 12/31/2012 3:55 109 12/31/2012 3:56 234 12/31/2012 3:57 267 12/31/2012 3:58 132 12/31/2012 3:59 265 12/31/2012 4:00 176 12/31/2012 4:01 155 12/31/2012 4:02 226 12/31/2012 4:03 189 12/31/2012 4:04 229 12/31/2012 4:05 136 12/31/2012 4:06 323 12/31/2012 4:07 254 12/31/2012 4:08 275 12/31/2012 4:09 208 12/31/2012 4:10 162 12/31/2012 4:11 253 12/31/2012 4:12 160 12/31/2012 4:13 176 12/31/2012 4:14 193 12/31/2012 4:15 340 12/31/2012 4:16 230 12/31/2012 4:17 175 12/31/2012 4:18 185 12/31/2012 4:19 247 12/31/2012 4:20 167 12/31/2012 4:21 168 12/31/2012 4:22 178 12/31/2012 4:23 199 12/31/2012 4:24 201 12/31/2012 4:25 178 12/31/2012 4:26 185 12/31/2012 4:27 212 12/31/2012 4:28 204 12/31/2012 4:29 214 12/31/2012 4:30 147 12/31/2012 4:31 209 12/31/2012 4:32 171 12/31/2012 4:33 115 12/31/2012 4:34 253 12/31/2012 4:35 111 12/31/2012 4:36 213 12/31/2012 4:37 259 12/31/2012 4:38 78 12/31/2012 4:39 209 12/31/2012 4:40 188 12/31/2012 4:41 204 12/31/2012 4:42 170 12/31/2012 4:43 269 12/31/2012 4:44 235 12/31/2012 4:45 214 12/31/2012 4:46 232 12/31/2012 4:47 172 12/31/2012 4:48 138 12/31/2012 4:49 161 12/31/2012 4:50 107 12/31/2012 4:51 114 12/31/2012 4:52 207 12/31/2012 4:53 146 12/31/2012 4:54 105 12/31/2012 4:55 200 12/31/2012 4:56 205 12/31/2012 4:57 199 12/31/2012 4:58 174 12/31/2012 4:59 208 12/31/2012 5:00 204 12/31/2012 5:01 247 12/31/2012 5:02 201 12/31/2012 5:03 163 12/31/2012 5:04 221 12/31/2012 5:05 275 12/31/2012 5:06 108 12/31/2012 5:07 172 12/31/2012 5:08 153
|
 |
|
| Reply #3
ZLM
    
1271 Posts |
Posted - 12/30/2012 : 17:33:13
|
On GMC-300, the minute data saving mode recording can last about 2 weeks. The Second data saving mode last about 14-16 hours.
The data from your log is high CPM, but this may not be true data. Sometimes the data is affected by downloading setup/process.
You may need to repeat downloading a few times t confirm the downloding has no problem.
Also, set the alarm ON. The default alarm CPM level is 100. So, you should hear the alarm if the CPM higher than 100.
|
 |
|
| Reply #4
fictitious1

4 Posts |
Posted - 12/30/2012 : 18:28:48
|
Here it is again on the heartbeat mode- I will again try to download the data from the device itself to make sure what it has recorded is correct. I actually turned the alarm off because it was, as you can see, going off the whole time this was above 100, and it was kinda scaring the kids wondering what was going on. I reset the alarm level to 300CPM, but turned it off as well. Its hooked up to my computer and I plan to just let it run for a few days. Wish I had this thing a year ago just to see what was going on. Gotta love radiation, can't see it can't smell it.
Date Time CPM 12/30/2012 22:16 38 12/30/2012 22:17 39 12/30/2012 22:18 123 12/30/2012 22:19 189 12/30/2012 22:20 213 12/30/2012 22:21 185 12/30/2012 22:22 85 12/30/2012 22:23 127 12/30/2012 22:24 218 12/30/2012 22:25 177 12/30/2012 22:26 201 12/30/2012 22:27 171 12/30/2012 22:28 104 12/30/2012 22:29 226 12/30/2012 22:30 71 12/30/2012 22:31 29 12/30/2012 22:32 43 12/30/2012 22:33 29 12/30/2012 22:34 27 12/30/2012 22:35 39 12/30/2012 22:36 30 12/30/2012 22:37 35 12/30/2012 22:38 27 12/30/2012 22:39 32 12/30/2012 22:40 43 12/30/2012 22:41 41 12/30/2012 22:42 38 12/30/2012 22:43 36 12/30/2012 22:44 31 12/30/2012 22:45 43 12/30/2012 22:46 73 12/30/2012 22:47 150 12/30/2012 22:48 135 12/30/2012 22:49 255 12/30/2012 22:50 217 12/30/2012 22:51 223 12/30/2012 22:52 207 12/30/2012 22:53 186 12/30/2012 22:54 168 12/30/2012 22:55 243 12/30/2012 22:56 239 12/30/2012 22:57 201 12/30/2012 22:58 195 12/30/2012 22:59 225 12/30/2012 23:00 227 12/30/2012 23:01 235 12/30/2012 23:02 135 12/30/2012 23:03 179 12/30/2012 23:04 200 12/30/2012 23:05 146 12/30/2012 23:06 104 12/30/2012 23:07 190 12/30/2012 23:08 154 12/30/2012 23:09 69 12/30/2012 23:10 144 12/30/2012 23:11 182 12/30/2012 23:12 201 12/30/2012 23:13 127 12/30/2012 23:14 191 12/30/2012 23:15 241 12/30/2012 23:16 140 12/30/2012 23:17 131 12/30/2012 23:18 126 12/30/2012 23:19 171 12/30/2012 23:20 250 12/30/2012 23:21 191 12/30/2012 23:22 192 12/30/2012 23:23 327 12/30/2012 23:24 136 12/30/2012 23:25 216 12/30/2012 23:26 150 12/30/2012 23:27 74 12/30/2012 23:28 69 12/30/2012 23:29 40 12/30/2012 23:30 28 12/30/2012 23:31 30 12/30/2012 23:32 37 12/30/2012 23:33 29 12/30/2012 23:34 40 12/30/2012 23:35 28 12/30/2012 23:36 30 12/30/2012 23:37 37 12/30/2012 23:38 32 12/30/2012 23:39 33 12/30/2012 23:40 21 12/30/2012 23:41 42 12/30/2012 23:42 30 12/30/2012 23:43 34 12/30/2012 23:44 51 12/30/2012 23:45 41 12/30/2012 23:46 37 12/30/2012 23:47 22 12/30/2012 23:48 26 12/30/2012 23:49 49 12/30/2012 23:50 48 12/30/2012 23:51 39 12/30/2012 23:52 28 12/30/2012 23:53 62 12/30/2012 23:54 122 12/30/2012 23:55 244 12/30/2012 23:56 135 12/30/2012 23:57 160 12/30/2012 23:58 134 12/30/2012 23:59 154 12/31/2012 0:00 85 12/31/2012 0:01 154 12/31/2012 0:02 256 12/31/2012 0:03 262 12/31/2012 0:04 160 12/31/2012 0:05 146 12/31/2012 0:06 145 12/31/2012 0:07 158 12/31/2012 0:08 167 12/31/2012 0:09 240 12/31/2012 0:10 256 12/31/2012 0:11 243 12/31/2012 0:12 176 12/31/2012 0:13 182 12/31/2012 0:14 181 12/31/2012 0:15 165 12/31/2012 0:16 229 12/31/2012 0:17 108 12/31/2012 0:18 177 12/31/2012 0:19 122 12/31/2012 0:20 120 12/31/2012 0:21 174 12/31/2012 0:22 168 12/31/2012 0:23 100 12/31/2012 0:24 137 12/31/2012 0:25 219 12/31/2012 0:26 126 12/31/2012 0:27 155 12/31/2012 0:28 133 12/31/2012 0:29 201 12/31/2012 0:30 178 12/31/2012 0:31 165 12/31/2012 0:32 230 12/31/2012 0:33 285 12/31/2012 0:34 226 12/31/2012 0:35 170 12/31/2012 0:36 97 12/31/2012 0:37 222 12/31/2012 0:38 154 12/31/2012 0:39 181 12/31/2012 0:40 146 12/31/2012 0:41 105 12/31/2012 0:42 93 12/31/2012 0:43 245 12/31/2012 0:44 198 12/31/2012 0:45 202 12/31/2012 0:46 256 12/31/2012 0:47 206 12/31/2012 0:48 143 12/31/2012 0:49 166 12/31/2012 0:50 187 12/31/2012 0:51 98 12/31/2012 0:52 138 12/31/2012 0:53 161 12/31/2012 0:54 196 12/31/2012 0:55 153 12/31/2012 0:56 171 12/31/2012 0:57 213 12/31/2012 0:58 140 12/31/2012 0:59 35 12/31/2012 1:00 39 12/31/2012 1:01 48 12/31/2012 1:02 154 12/31/2012 1:03 180 12/31/2012 1:04 153 12/31/2012 1:05 296 12/31/2012 1:06 142 12/31/2012 1:07 200 12/31/2012 1:08 97 12/31/2012 1:09 160 12/31/2012 1:10 217 12/31/2012 1:11 169 12/31/2012 1:12 178 12/31/2012 1:13 152 12/31/2012 1:14 101 12/31/2012 1:15 193 12/31/2012 1:16 170 12/31/2012 1:17 258 12/31/2012 1:18 178 12/31/2012 1:19 156 12/31/2012 1:20 193 12/31/2012 1:21 138 12/31/2012 1:22 231 12/31/2012 1:23 137 12/31/2012 1:24 125 12/31/2012 1:25 219 12/31/2012 1:26 213 12/31/2012 1:27 291 12/31/2012 1:28 179 12/31/2012 1:29 168 12/31/2012 1:30 175 12/31/2012 1:31 133 12/31/2012 1:32 171 12/31/2012 1:33 124 12/31/2012 1:34 211 12/31/2012 1:35 127 12/31/2012 1:36 249 12/31/2012 1:37 164 12/31/2012 1:38 216 12/31/2012 1:39 161 12/31/2012 1:40 278 12/31/2012 1:41 181 12/31/2012 1:42 184 12/31/2012 1:43 168 12/31/2012 1:44 227 12/31/2012 1:45 238 12/31/2012 1:46 148 12/31/2012 1:47 35 12/31/2012 1:48 49 12/31/2012 1:49 39 12/31/2012 1:50 45 12/31/2012 1:51 26 12/31/2012 1:52 30 12/31/2012 1:53 24 12/31/2012 1:54 46 12/31/2012 1:55 35 12/31/2012 1:56 115 12/31/2012 1:57 123 12/31/2012 1:58 114 12/31/2012 1:59 148 12/31/2012 2:00 193 12/31/2012 2:01 235 12/31/2012 2:02 237 12/31/2012 2:03 192 12/31/2012 2:04 229 12/31/2012 2:05 198 12/31/2012 2:06 116 12/31/2012 2:07 182 12/31/2012 2:08 157 12/31/2012 2:09 262 12/31/2012 2:10 99 12/31/2012 2:11 214 12/31/2012 2:12 177 12/31/2012 2:13 133 12/31/2012 2:14 192 12/31/2012 2:15 199 12/31/2012 2:16 190 12/31/2012 2:17 195 12/31/2012 2:18 251 12/31/2012 2:19 183 12/31/2012 2:20 154 12/31/2012 2:21 178 12/31/2012 2:22 255 12/31/2012 2:23 165 12/31/2012 2:24 309 12/31/2012 2:25 266 12/31/2012 2:26 187 12/31/2012 2:27 251 12/31/2012 2:28 191 12/31/2012 2:29 178 12/31/2012 2:30 105 12/31/2012 2:31 273 12/31/2012 2:32 140 12/31/2012 2:33 178 12/31/2012 2:34 230 12/31/2012 2:35 205 12/31/2012 2:36 186 12/31/2012 2:37 191 12/31/2012 2:38 187 12/31/2012 2:39 210 12/31/2012 2:40 210 12/31/2012 2:41 204 12/31/2012 2:42 222 12/31/2012 2:43 210 12/31/2012 2:44 180 12/31/2012 2:45 239 12/31/2012 2:46 162 12/31/2012 2:47 286 12/31/2012 2:48 230 12/31/2012 2:49 212 12/31/2012 2:50 198 12/31/2012 2:51 148 12/31/2012 2:52 186 12/31/2012 2:53 182 12/31/2012 2:54 225 12/31/2012 2:55 184 12/31/2012 2:56 209 12/31/2012 2:57 183 12/31/2012 2:58 227 12/31/2012 2:59 352 12/31/2012 3:00 190 12/31/2012 3:01 215 12/31/2012 3:02 228 12/31/2012 3:03 229 12/31/2012 3:04 161 12/31/2012 3:05 212 12/31/2012 3:06 223 12/31/2012 3:07 281 12/31/2012 3:08 103 12/31/2012 3:09 306 12/31/2012 3:10 303 12/31/2012 3:11 151 12/31/2012 3:12 209 12/31/2012 3:13 267 12/31/2012 3:14 204 12/31/2012 3:15 263 12/31/2012 3:16 174 12/31/2012 3:17 203 12/31/2012 3:18 218 12/31/2012 3:19 200 12/31/2012 3:20 210 12/31/2012 3:21 156 12/31/2012 3:22 115 12/31/2012 3:23 256 12/31/2012 3:24 205 12/31/2012 3:25 239 12/31/2012 3:26 192 12/31/2012 3:27 256 12/31/2012 3:28 246 12/31/2012 3:29 137 12/31/2012 3:30 209 12/31/2012 3:31 145 12/31/2012 3:32 204 12/31/2012 3:33 149 12/31/2012 3:34 213 12/31/2012 3:35 230 12/31/2012 3:36 147 12/31/2012 3:37 272 12/31/2012 3:38 187 12/31/2012 3:39 198 12/31/2012 3:40 129 12/31/2012 3:41 213 12/31/2012 3:42 198 12/31/2012 3:43 291 12/31/2012 3:44 275 12/31/2012 3:45 345 12/31/2012 3:46 266 12/31/2012 3:47 143 12/31/2012 3:48 142 12/31/2012 3:49 224 12/31/2012 3:50 246 12/31/2012 3:51 200 12/31/2012 3:52 170 12/31/2012 3:53 276 12/31/2012 3:54 212 12/31/2012 3:55 109 12/31/2012 3:56 234 12/31/2012 3:57 267 12/31/2012 3:58 132 12/31/2012 3:59 265 12/31/2012 4:00 176 12/31/2012 4:01 155 12/31/2012 4:02 226 12/31/2012 4:03 189 12/31/2012 4:04 229 12/31/2012 4:05 136 12/31/2012 4:06 323 12/31/2012 4:07 254 12/31/2012 4:08 275 12/31/2012 4:09 208 12/31/2012 4:10 162 12/31/2012 4:11 253 12/31/2012 4:12 160 12/31/2012 4:13 176 12/31/2012 4:14 193 12/31/2012 4:15 340 12/31/2012 4:16 230 12/31/2012 4:17 175 12/31/2012 4:18 185 12/31/2012 4:19 247 12/31/2012 4:20 167 12/31/2012 4:21 168 12/31/2012 4:22 178 12/31/2012 4:23 199 12/31/2012 4:24 201 12/31/2012 4:25 178 12/31/2012 4:26 185 12/31/2012 4:27 212 12/31/2012 4:28 204 12/31/2012 4:29 214 12/31/2012 4:30 147 12/31/2012 4:31 209 12/31/2012 4:32 171 12/31/2012 4:33 115 12/31/2012 4:34 253 12/31/2012 4:35 111 12/31/2012 4:36 213 12/31/2012 4:37 259 12/31/2012 4:38 78 12/31/2012 4:39 209 12/31/2012 4:40 188 12/31/2012 4:41 204 12/31/2012 4:42 170 12/31/2012 4:43 269 12/31/2012 4:44 235 12/31/2012 4:45 214 12/31/2012 4:46 232 12/31/2012 4:47 172 12/31/2012 4:48 138 12/31/2012 4:49 161 12/31/2012 4:50 107 12/31/2012 4:51 114 12/31/2012 4:52 207 12/31/2012 4:53 146 12/31/2012 4:54 105 12/31/2012 4:55 200 12/31/2012 4:56 205 12/31/2012 4:57 199 12/31/2012 4:58 174 12/31/2012 4:59 208 12/31/2012 5:00 204 12/31/2012 5:01 247 12/31/2012 5:02 201 12/31/2012 5:03 163 12/31/2012 5:04 221 12/31/2012 5:05 275 12/31/2012 5:06 108 12/31/2012 5:07 172 12/31/2012 5:08 153 12/31/2012 5:09 208 12/31/2012 5:10 154 12/31/2012 5:11 126 12/31/2012 5:12 109 12/31/2012 5:13 222 12/31/2012 5:14 151 12/31/2012 5:15 209 12/31/2012 5:16 142 12/31/2012 5:17 174 12/31/2012 5:18 154 12/31/2012 5:19 42 12/31/2012 5:20 103 12/31/2012 5:21 102 12/31/2012 5:22 134 12/31/2012 5:23 101 12/31/2012 5:24 117 12/31/2012 5:25 129 12/31/2012 5:26 49 12/31/2012 5:27 26 12/31/2012 5:28 18 12/31/2012 5:29 25 12/31/2012 5:30 25 12/31/2012 5:31 31 12/31/2012 5:32 32 12/31/2012 5:33 29 12/31/2012 5:34 39 12/31/2012 5:35 28 12/31/2012 5:36 49 12/31/2012 5:37 28 12/31/2012 5:38 41 12/31/2012 5:39 27 12/31/2012 5:40 26 12/31/2012 5:41 33 12/31/2012 5:42 20 12/31/2012 5:43 38 12/31/2012 5:44 16 12/31/2012 5:45 35 12/31/2012 5:46 33 12/31/2012 5:47 38 12/31/2012 5:48 38 12/31/2012 5:49 21 12/31/2012 5:50 25 12/31/2012 5:51 36 12/31/2012 5:52 21 12/31/2012 5:53 32 12/31/2012 5:54 34 12/31/2012 5:55 33 12/31/2012 5:56 34 12/31/2012 5:57 38 12/31/2012 5:58 30 12/31/2012 5:59 28 12/31/2012 6:00 27 12/31/2012 6:01 17 12/31/2012 6:02 18 12/31/2012 6:03 32 12/31/2012 6:04 34 12/31/2012 6:05 24 12/31/2012 6:06 33 12/31/2012 6:07 36 12/31/2012 6:08 33 12/31/2012 6:09 39 12/31/2012 6:10 31 12/31/2012 6:11 11 12/31/2012 6:12 31 12/31/2012 6:13 22 12/31/2012 6:14 24 12/31/2012 6:15 18 12/31/2012 6:16 22 12/31/2012 6:17 36 12/31/2012 6:18 27 12/31/2012 6:19 31 12/31/2012 6:20 37 12/31/2012 6:21 27 12/31/2012 6:22 24 12/31/2012 6:23 16 12/31/2012 6:24 15 12/31/2012 6:25 21 12/31/2012 6:26 23 12/31/2012 6:27 28 12/31/2012 6:28 21 12/31/2012 6:29 18 12/31/2012 6:30 26 12/31/2012 6:31 40 12/31/2012 6:32 22 12/31/2012 6:33 28 12/31/2012 6:34 35 12/31/2012 6:35 15 12/31/2012 6:36 15 12/31/2012 6:37 29 12/31/2012 6:38 33 12/31/2012 6:39 36 12/31/2012 6:40 38 12/31/2012 6:41 30 12/31/2012 6:42 18 12/31/2012 6:43 32 12/31/2012 6:44 30 12/31/2012 6:45 40 12/31/2012 6:46 24 12/31/2012 6:47 35 12/31/2012 6:48 25 12/31/2012 6:49 28 12/31/2012 6:50 34 12/31/2012 6:51 32 12/31/2012 6:52 20 12/31/2012 6:53 16 12/31/2012 6:54 17 12/31/2012 6:55 35 12/31/2012 6:56 22 12/31/2012 6:57 16 12/31/2012 6:58 25 12/31/2012 6:59 8 12/31/2012 7:00 17
|
 |
|
| Reply #5
ZLM
    
1271 Posts |
Posted - 12/30/2012 : 18:40:50
|
Looks like the internal geiger tube may be involved.
Open the case and try to turn/move the tube little bit and check if it has any contact problem.
If the problem still there, please write to support@gqelectronicsllc.com for further assistant.
|
 |
|
| Reply #6
fictitious1

4 Posts |
Posted - 12/30/2012 : 20:53:22
|
Hello,
Thanks for the advice. I opened it up and saw a bit of corrosion on the contacts at the end of the tube where it makes contact with the clip. I love tinkering with things anyway but would not have opened it on my own without it being suggested . I cleaned it up as best as I could with some contact cleaner, rolling the tube a little bit at a time and cleaning it. I did not want to clip the ties that were holding the tube to the board(is that ok though? and then use your own) so I just did what I could by trying to roll it around slowly. I'll record for another day or so and see how things go. Cleaned up ok I think.
Thanks again. |
 |
|
| |