atomic.dave

USA
65 Posts

Posted - 01/14/2012 :  22:27:04  Show Profile  Reply with Quote
How do you calibrate the GMC280?
Reply #1

ZLM

1271 Posts

Posted - 01/15/2012 :  01:37:37  Show Profile  Reply with Quote
GMC-080, GMC-280 CPM reading does not need a special calibration, as long as the tube voltage is set correctly. The on board hardwae take care the accuracy of CPM.

If you want uSv/h or mR/h reading correctly, the factory default value is for M4011 tube or similar tube such as SBM-20. You do not need to do calibration on those those two tubes. Unless you found the value is offset too much.

Other tubes need a reference source with a known uSv/h. Then place the source with GMC-280 to get CPM, and input this CPM and the known uSv/h in calibration menu. It accepts up to three reference values.


The first calibration data CPM should be always background CPM, then the uSv/h value normally is 0.05 uSv/h. This 0.05 uSv/h value is the typical background radiation dosage value. If you location is near rocky mont, then the value may be higher.



Go to Top of Page