GQ Electronics Technical Support Forum Active Users: / Visits Today:
Highest Active Users:
GQ Electronics Technical Support Forum
Home | Profile | Register | Active Topics | Members | Search | FAQ
Username:
Password:
Save Password
Forgot your Password?

 All Forums
 GQ Electronics Forums
 2.GQ Geiger Muller Counter
 problems getting data from GM-300E

Note: You must be registered in order to post a reply.
To register, click here. Registration is FREE!

Screensize:
UserName:
Password:
Format Mode:
Format: BoldItalicizedUnderlineStrikethrough Align LeftCenteredAlign Right Horizontal Rule Insert HyperlinkInsert EmailInsert Image Insert CodeInsert QuoteInsert List Spell Checker
   
Message:

* HTML is OFF
* Forum Code is ON
Smilies
Smile [:)] Big Smile [:D] Cool [8D] Blush [:I]
Tongue [:P] Evil [):] Wink [;)] Clown [:o)]
Black Eye [B)] Eight Ball [8] Frown [:(] Shy [8)]
Shocked [:0] Angry [:(!] Dead [xx(] Sleepy [|)]
Kisses [:X] Approve [^] Disapprove [V] Question [?]

   Insert an Image File
Check here to include your profile signature.
    

T O P I C    R E V I E W
robsmoto Posted - 09/20/2013 : 07:44:49
I have several older GM-300 units and a newer GM-300E. My problem is with the GM-300E, specifically the inability to extract the data. I am using the GM data viewer ver 2.03. The GM-300 data transfers from the detector to the computer just fine. The GM-300E appears to connect ok (e.g., same message com3 opened successfully), but when downloading I get a message "There are som unsynched communication. Check the unit communication baud setting. You may need to restart/reconnect the unit and download again."

I have tried several baud settings (i.e., 9600, 8 N 1 and 57600 8 N 1) which work for the GM-300, but none work for the GM-300E. I've tried using the device manager and the GM data viewer to get consistent baud settings - the transfers with the GM-300 occur just fine, but not with the GM-300E.

Help!
2   L A T E S T    R E P L I E S    (Newest First)
robsmoto Posted - 09/21/2013 : 12:39:31
There must have been some corrupt data. I did the factory reset as suggested and did a brief test. With the com port (#3) set at 57600 8N1 I was able to successfully extract the data. Thank you!

PS These devices are a great value.
ZLM Posted - 09/21/2013 : 07:05:28
The GMC-300E uses fix baud rate: 57600 BPS. 9600 does not work.

If you always get this message, then your GMC-300E internal flash memory may have some corrupted data. Do a factory reset will erase the flash data and restart again. That should solve the problem.