Posted - 05/17/2013 : 00:29:47 Hi,
I am the new owner of a GMC-300. The unit work nicely. But when I test samples with high activity the unit take a very long time (a dozen of minutes) to go back to normal values. Is this related to the Geiger tube or hardware, or to the leveraging algorithm for CPM ? What is the integration time for CPM ? The last minute ? The last x minutes ?
I tested this :
- Background is 20 CPM - GMC is exposed to fiestaware peaking at 4000 CPM for 1 minute. Then Fiestaware is removed. - After 6 minutes the GMC still display 140 CPM, and not the real background value which oscillate between 15 and 30 CPM.
I don't use data recording only realtime display. Is there anything that can be done ?
Also Geiger tubes have some dead time, what is the CPM value where CPM display start to be not linear ?
Thank you very much :) |
2 L A T E S T R E P L I E S (Newest First) |
harlan |
Posted - 06/05/2013 : 11:02:21 I had the same problem with my GMC-300 when I went radiation antiquing in my mall that sell old things. I thought the whole shop was radioactive because the GMC-300 got readings over 5000 CPM. I bought the fiestaware cup and saucer and then left the store. The next day, I went back and the readings were normal. So now I know to leave the area and turn off the unit, then come back and check it again. |
ZLM |
Posted - 05/17/2013 : 00:44:08 It is related to the tube. You can turn it off and then on after a high count measurement. That will help to back to normal faster. The CPM is a realtime and real count, so it dies not have linear issue. Only when CPM converting to dose unit has linear issue. But I do not have such information now. |